Summary of symbols
The software uses a grouping scheme for categorizing variables. Variables displayed in a Measurements grid or logged in a data file will be labeled with a group name and a display label. Some groups (e.g., Meas) are available all the time, and some (e.g., Aquatic) are available only in certain configurations, determined by the attached chamber, and the settings on the Start Up > Chamber Setup and/or Peripheral Setup pages.
While any of these variables can be displayed on the Measurements grid, most are not there by default. However, a simple way to see what groups are available and to check current values is Start Up > Data Dictionary.
What follows are tables for each of the approximately 40 defined groups. Each entry shows the Display Label (the name that appears on the interface, or in the log file), its units, a description, and the internal name. The latter is used to reference the variable from a background program (BP), for example.
Aquatic
Table 4‑4 . Aquatic chamber configuration and settings.
| Fan On |
|
Chamber circulation fan |
FanOn |
| Pump On |
|
Aquatic sample air flow |
PumpOn |
| Pump Speed |
V |
Operating range 2 to 5 V |
PumpV |
| Bubble effect |
|
Bubble effect on tau_w |
bubble |
| Blue adjust |
|
Quantum sensor cal adjust, blue |
qc_b |
| Red adjust |
|
Quantum sensor cal adjust, red |
qc_r |
| tau_wb |
|
Chamber clear water transmittance, blue |
tw_b |
| tau_wr |
|
Chamber clear water transmittance, red |
tw_r |
AquaticQ
Table 4‑5. Aquatic chamber light related measurements and computations.
| Qabs |
µmol m-2 s-1 |
Flux absorbed by algae |
Qabs |
| Qabs* |
µmol s-1 |
Absorbed quantum flow (Qabs X window area) |
Qabs* |
| Qavg |
µmol m-2 s-1 |
Average in-sample irradiance |
Qavg |
| Qin |
µmol m-2 s-1 |
Flux incident on aquatic sample |
Qin |
| Qout |
µmol m-2 s-1 |
Flux leaving aquatic sample |
Qout |
| tau_s |
|
Measured sample transmittance |
tau_s |
| tau_w |
|
Transmittance of water (adjusted for color) |
tau_w |
Auxiliary
Table 4‑6. Settings and measurements associated with the Console USER I/O Port.
| ADC_CH1 |
V |
Spare channel 1 |
ADC_CH1 |
| ADC_CH2 |
V |
Spare channel 2 |
ADC_CH2 |
| ADC_CH3 |
V |
Spare channel 3 |
ADC_CH3 |
| ADC_CH4 |
V |
Spare channel 4 |
ADC_CH4 |
| ADC_CH5 |
V |
Spare channel 5 |
ADC_CH5 |
| ADC_CH6 |
V |
Spare channel 6 |
ADC_CH6 |
| ADC_CH7 |
V |
Spare channel 7 |
ADC_CH7 |
| ADC_CH8 |
V |
Spare channel 8 |
ADC_CH8 |
| AuxPower |
V |
Auxiliary power setting (leaf chamber) |
AuxPower |
| GPIO |
|
GPIO Status channels 8 through 1 |
GPIO |
| GPIO_dir |
|
GPIO Direction channels 8 through 1 |
GPIO_dir |
| ch1_pullup |
|
Pullup resistor for ADC_CH1 |
adc_ch1_pullup |
| DAC_1 |
V |
D/A setting (-5 to +5V) |
dac1 |
| DAC_2 |
V |
D/A setting (-5 to +5V) |
dac2 |
| DAC_3 |
V |
D/A setting (-5 to +5V) |
dac3 |
| DAC_4 |
V |
D/A setting (-5 to +5V) |
dac4 |
| excit_5v |
|
5 V excitation line |
excitation_5v |
| power_12v |
|
12 V power |
power_12v |
| power_5v |
|
5 V power |
power_5v |
ChambConst
Table 4‑7. Constants associated with the currently selected chamber.
| Aperture |
cm2 |
Attached chamber aperture |
Aperture |
| Chamber |
|
Attached chamber ID |
Chamber |
| blc_Po |
kPa |
Reference pressure for blc calculation |
blc_Po |
| blc_a |
|
Boundary layer coefficient ’a’ |
blc_a |
| blc_b |
|
Boundary layer coefficient ’b’ |
blc_b |
| blc_c |
|
Boundary layer coefficient ’c’ |
blc_c |
| blc_d |
|
Boundary layer coefficient ’d’ |
blc_d |
| blc_e |
|
Boundary layer coefficient ’e’ |
blc_e |
| blc_maxS |
cm2 |
Maximum area for blc calculation |
blc_maxS |
| blc_minS |
cm2 |
Minimum area for blc calculation |
blc_minS |
ConsoleLS
Table 4‑8. Settings associated with the light source connected to the console (if any).
| Pc |
|
Photodiode count |
Pc |
| Q |
µmol m-2 s-1 |
Light Source output |
Q |
| Tled |
C |
LED tile temperature |
TLEDTile |
| f_red |
|
Fraction red |
f_1 |
| f_green |
|
Fraction green |
f_2 |
| f_blue |
|
Fraction blue |
f_3 |
| f_white |
|
Fraction white |
f_4 |
ConStatus
Table 4‑9. Console status values.
| CO2_f |
V |
CO2 Injector flow |
CO2_FLOW |
| CO2_f_s |
V |
CO2 Injector status |
CO2_FLOW_STATUS |
| Pump_f |
V |
Pump flow |
PUMP_FLOW |
| Pump_f_s |
V |
Pump status |
PUMP_FLOW_STATUS |
| Pump_p |
V |
Pump pressure flow |
PUMP_PRESSURE |
| Pump_p_s |
V |
Pump pressure status |
PUMP_PRESSURE_STATUS |
| V_system |
V |
System voltage |
SYSTEM_VOLTAGE |
| Tboard |
C |
Console board temperature |
TEMPERATURE |
Const
These constants are associated with the Gas Exchange computations, and depend on how the Start Up > Chamber Setup page is configured.
Table 4‑10. Constants for typical gas exchange measurements.
| Custom |
mol m-2 s-1 |
1-sided BLC |
CustomBLC |
| Geometry |
|
Used for boundary layer conductance (BLC) |
Geometry |
| K |
|
Stomatal ratio |
K |
| S |
cm2 |
Leaf area |
S |
Table 4‑11. Constants for the aquatic chamber.
| Dcell |
cells ml-1 |
Algal cell density |
Dcell |
| Dchl |
µg ml-1 |
Chlorophyll density |
Dchl |
| Dmass |
mg ml-1 |
Algal mass density |
Dmass |
| Volume |
ml |
Sample volume |
V |
Dynamic
Table 4‑12. Computations associated with dynamic assimilation and transpiration.
| Adyn |
µmol m-2 s-1 |
Assimilation via dynamic method |
Adyn |
| Crd |
µmol mol-1 |
CO2 ref, dry equivalent, time lag adjusted |
Crd |
| Csd |
µmol mol-1 |
CO2 sample, dry equivalent, from regression |
Csd |
| Edyn |
mmol m-2 s-1 |
Transpiration via dynamic method |
Edyn |
| Hr |
mmol mol-1 |
H2O ref, time lag adjusted |
Hr |
| Hs |
mmol mol-1 |
H2O sample, from regression |
Hs |
| αVc |
cm3 |
Effective CO2 Volume |
Volume |
| αVh |
cm3 |
Effective H2O Volume |
VolumeH |
| dCsd/dt |
µmol mol-1 s-1 |
Rate of change of Csd |
dCsd/dt |
| dHs/dt |
mmol mol-1 s-1 |
Rate of change of Hs |
dHs/dt |
DynConfig
Table 4‑13. The dynamic configuration settings.
| Min CO2 Slope |
µmol mol-1 s-1 |
Minimum usable slope for CO2 |
co2_slope_threshold |
| Min H2O Slope |
mmol mol-1 s-1 |
Minimum usable slope for H2O |
h2o_slope_threshold |
FastKntcs
Table 4‑14. Computations from the latest induction flash.
| Duration |
ms |
Flash duration |
DURATION |
| ID |
|
ID of this INDUCTION flash |
EVENT_ID |
| F1 |
|
First inflection |
F1 |
| F2 |
|
Second inflection |
F2 |
| Fmax |
|
Max F during flash |
FMAX |
| Fo |
|
F before flash |
Fo |
| T@F1 |
s |
Time of F1 |
T@F1 |
| T@F2 |
s |
Time of F2 |
T@F2 |
| T@Fmax |
s |
Time of Fmax |
T@FMAX |
| T@HIR |
s |
Time of half the initial rise |
T@HIR |
FLR
Table 4‑15. Fluorometer computations, updated after every flash and/or dark event.
| 1-qL |
|
1-qL |
1-qL |
| Flux*_fs |
µmol s-1 |
Flux at Fs |
A |
| Flux*_d |
µmol s-1 |
Dark assimilation rate before Fm |
Adark |
| DarkAdaptedID |
|
ID of latest dark-adapted flash |
DarkAdaptedID |
| DarkPulseID |
|
ID of latest dark pulse |
DarkPulseID |
| ETR |
µmol s-1 |
Electron transport rate |
ETR |
| Fm |
|
Maximal F, dark adapted |
Fm |
| Fmin |
|
Measured minimal F, light adapted |
Fmin |
| Fm’ |
|
Maximal F, light adapted |
Fmp |
| Fo |
|
Minimal F, dark adapted |
Fo |
| Fo’ |
|
Minimal F, light adapted |
Fop |
| Fs |
|
Steady state F |
Fs |
| Fv/Fm |
|
1-Fo/Fm |
Fv/Fm |
| Fv’/Fm’ |
|
1-Fo’/Fm’ |
Fvp/Fmp |
| LightAdaptedID |
|
ID of latest light-adapted flash |
LightAdaptedID |
| NPQ |
|
(Fm-Fm’)/Fm’ |
NPQ |
| PS2/1 |
|
Ratio PS2 to PS1 |
PS2/1 |
| PhiCO2 |
µmol µmol-1 |
(A-Ao)/Qabs |
PhiCO2 |
| PhiPS2 |
|
1-Fs/Fm’ |
PhiPS2 |
| Qabs*_fs |
µmol s-1 |
Quanta/s absorbed at Fs |
Qabs |
| Qmax |
|
QMax for latest light adapted flash |
Qmax |
| Qmax_d |
|
QMax for latest dark adapted flash |
Qmax_d |
| alt. Fo’ |
|
Alternative Fo’ = Fo/(Fv/Fm+Fo/Fm’) |
aFop |
| qL |
|
qP Fo’/Fs |
qL |
| qN |
|
(Fm-Fm’)/(Fm-Fo’) |
qN |
| qN_Fo |
|
(Fm-Fm’)/(Fm-Fo) |
qN_Fo |
| qP |
|
(Fm’-Fs)/(Fm’-Fo’) |
qP |
| qP_Fo |
|
(Fm’-Fs)/(Fm’-Fo) |
qP_Fo |
FlrLS
Table 4‑16. Status and readings associated with the fluorometer light source.
| f_blue |
|
Fraction blue |
BLUEFRAC |
| F_dc |
|
DC fluorescence |
DC |
| f_farred |
|
Fraction farred |
FARREDFRAC |
| F |
|
Demodulated fluorescence |
FLUOR |
| Pc |
|
Photodiode count |
PC |
| Q |
µmol m-2 s-1 |
Fluorometer photon flux density |
PFD |
| Q_blue |
µmol m-2 s-1 |
Blue actinic contribution to Q |
Q_blue |
| Q_farred |
µmol m-2 s-1 |
Far red contribution to Q |
Q_farred |
| Q_red |
µmol m-2 s-1 |
Red actinic contribution to Q |
Q_red |
| f_red |
|
Fraction red |
REDFRAC |
| Q_modavg |
µmol m-2 s-1 |
Modulated contribution to Q |
REDMODAVG |
| TDigital |
°C |
Digital board temperature |
TDigital |
| TDrive |
°C |
Driver temperature |
TDrive |
| Tled |
°C |
LED tile temperature |
TLEDTile |
| TPwrSpy |
°C |
Power supply temperature |
TPowerSupply |
| TPreamp |
°C |
Preamp board temperature |
TPreamp |
| TSPF |
s |
Time since previous flash |
TSPF |
| state |
|
0=Normal, 1=Rect, 2=MPF, 3=Ind, 4=Dark |
state |
FlrStats
Table 4‑17. Statistics on the fluorometer’s modulated fluorescence signal.
| F_avg |
|
Average F |
F_avg |
| F_dc_avg |
|
Average F_dc |
F_dc_avg |
| dF/dt |
min-1 |
Rate of change of F |
dF/dt |
| dF_dc/dt |
min-1 |
Rate of change of F_dc |
dF_dc/dt |
| period |
s |
Period for stats |
period |
GasEx
Gas exchange computations. These depend on how the Start Up > Chamber Setup page is configured.
Table 4‑18. Example for normal gas exchange computations.
| A |
µmol m-2 s-1 |
Assimilation rate |
A |
| Ca |
µmol mol-1 |
Ambient (to leaf) CO2 |
Ca |
| Ci |
µmol mol-1 |
Intercellular CO2 |
Ci |
| E |
mol m-2 s-1 |
Transpiration rate |
E |
| EBSum |
W m-2 |
Sum of energy balance components |
EBSum |
| Emm |
mmol m-2 s-1 |
Transpiration rate |
Emm |
| LatHFlux |
W m-2 |
Latent Heat Flux |
LatHFlux |
| NetTherm |
W m-2 |
Net Thermal |
NetTherm |
| Pca |
Pa |
Ambient (to leaf) CO2 |
Pca |
| Pci |
Pa |
Intercellular CO2 |
Pci |
| RHcham |
% |
Relative humidity in the chamber |
RHcham |
| Rabs |
W m-2 |
Absorbed short wave radiation |
Rabs |
| SVPcham |
kPa |
Saturation vapor pressure in the chamber |
SVPcham |
| SVPleaf |
kPa |
Saturation vapor pressure at leaf temp |
SVPleaf |
| SenHFlux |
W m-2 |
Sensible Heat Flux |
SenHFlux |
| TIME |
s |
Seconds since Jan 1970 |
TIME |
| TleafCnd |
°C |
Leaf temperature used for gas exchange computations |
TleafCnd |
| TleafEB |
°C |
Leaf temperature from energy balance |
TleafEB |
| VPDleaf |
kPa |
Vapor pressure deficit at leaf temp |
VPDleaf |
| VPcham |
kPa |
Vapor pressure in the chamber |
VPcham |
| gbw |
mol m-2 s-1 |
Boundary layer conductance to water vapor |
gbw |
| gsw |
mol m-2 s-1 |
Stomatal conductance to water vapor |
gsw |
| gtc |
mol m-2 s-1 |
Total conductance to CO2 |
gtc |
| gtw |
mol m-2 s-1 |
Total conductance to water vapor |
gtw |
Table 4‑19. Example for the Aquatic Chamber.
| Flux* |
µmol s-1 |
Carbon flow |
A |
| Flux_cell |
µmol cell-1 s-1 |
Algal cell flux |
Acell |
| Flux_chl |
µmol µg-1 s-1 |
Chlorophyll based flux |
Achl |
| Flux_mass |
µmol mg-1 s-1 |
Algal mass flux |
Amass |
| TIME |
s |
Seconds since Jan 1970 |
TIME |
| Taq |
°C |
Aquatic chamber temp |
Taq |
| pH |
|
Aquatic chamber pH |
pH |
| pH Signal |
V |
Aquatic chamber pH probe signal |
pH_V |
HeadLS
Table 4‑20. Settings associated with the light source connected to the head (if any).
| Pc |
|
Photodiode count |
Pc |
| Q |
µmol m-2 s-1 |
Light Source output |
Q |
| Tled |
°C |
LED tile temperature |
TLEDTile |
| f_red |
|
Fraction red |
f_1 |
| f_green |
|
Fraction green |
f_2 |
| f_blue |
|
Fraction blue |
f_3 |
| f_white |
|
Fraction white |
f_4 |
Leak
Table 4‑21. Computations associated with the leak correction.
| Fan |
µmol s-1 |
Estimated mixing fan flow |
Fan |
| Leak |
µmol s-1 |
Leakage from chamber |
Leak |
| CorrFact |
|
Leak correction factor |
LeakK |
| CorrFactPct |
% |
Leak correction factor as percent |
LeakPE |
| LeakPct |
% |
Leakage as percent of flow |
LeakPct |
LeakConst
Table 4‑22. Constants associated with the leak correction.
| Fs_meas |
µmol s-1 |
Flow_s_adjust table, vector of measured values |
Fs_meas |
| Fs_true |
µmol s-1 |
Flow_s adjust table, vector of true values |
Fs_true |
| fan_a |
µmol s-1 |
Fanflow cal coeff |
fan_a |
| fan_b |
µmol s-1/rpm |
Fanflow cal coeff |
fan_b |
| fan_c |
1/(rpm) |
Fanflow cal coeff |
fan_c |
| fan_d |
K/kPa |
Fanflow cal coeff |
fan_d |
| leak_wt |
|
Leak weighting factor |
leak_wt |
LeafQ
Table 4‑23. Light computations associated with a leaf in the chamber.
| Qabs |
µmol m-2 s-1 |
PPFD absorbed by the leaf |
Qabs |
| Qin |
µmol m-2 s-1 |
PPFD incident on the leaf |
Qin |
| alpha |
|
Weighted leaf absorption |
alpha |
| convert |
J/µmol |
Weighted conversion from PPFD to W/m2 |
convert |
LQConst
Table 4‑24. Leaf related constants associated with the LeafQ computations.
| Ambient |
|
Ambient light identifier |
Ambient |
| Leaf |
|
Leaf type identifier |
Leaf |
| abs_ambient |
|
Leaf PAR absorptance for ambient |
abs_ambient |
| abs_blueFlr |
|
Leaf PAR absorptance for blueFlr |
abs_blueFlr |
| abs_blueLED |
|
Leaf PAR absorptance for blueLED |
abs_blueLED |
| abs_greenLED |
|
Leaf PAR absorptance for greenLED |
abs_greenLED |
| abs_redFlr |
|
Leaf PAR absorptance for redFlr |
abs_redFlr |
| abs_redLED |
|
Leaf PAR absorptance for redLED |
abs_redLED |
| abs_whiteLED |
|
Leaf PAR absorptance for whiteLED |
abs_whiteLED |
| k_ambient |
J/μmol |
Converts Qin to Rabs for ambient |
k_ambient |
| k_blueFlr |
J/μmol |
Converts Qin to Rabs for blueFlr |
k_blueFlr |
| k_blueLED |
J/μmol |
Converts Qin to Rabs for blueLED |
k_blueLED |
| k_greenLED |
J/μmol |
Converts Qin to Rabs for greenLED |
k_greenLED |
| k_redFlr |
J/μmol |
Converts Qin to Rabs for redFlr |
k_redFlr |
| k_redLED |
J/μmol |
Converts Qin to Rabs for redLED |
k_redLED |
| k_whiteLED |
J/μmol |
Converts Qin to Rabs for whiteLED |
k_whiteLED |
LTConst
Table 4‑25. Constants associated with determining the temperature used for stomatal conductance.
| deltaTw |
°C |
Wall-air temp difference |
deltaTw |
| fT1 |
|
Fraction Tleaf |
fT1 |
| fT2 |
|
Fraction Tleaf2 |
fT2 |
| fT3 |
|
Fraction TleafEB |
fTeb |
MchEvent
Table 4‑26.Results associated with the latest point match event.
| co2_adj |
µmol mol-1 |
Pre-match error: CO2_r - CO2_s at last match. Ideally 0. |
co2_adj |
| co2_at |
µmol mol-1 |
CO2 at last match |
co2_at |
| co2_cv |
% |
CV of last CO2 match |
co2_cv |
| co2_match |
µmol mol-1 |
Match correction: CO2_r - CO2_a at last pt. match. |
co2_match |
| co2_t |
s |
Last CO2 match time |
co2_t |
| count |
|
Match count number |
count |
| h2o_adj |
mmol mol-1 |
Pre-match error: H2O_r - H2O_s at last match. Ideally 0. |
h2o_adj |
| h2o_at |
mmol mol-1 |
H2O at last match |
h2o_at |
| h2o_cv |
% |
CV of last H2O match |
h2o_cv |
| h2o_match |
mmol mol-1 |
Match correction: H2O_r - H2O_a at last pt. match. |
h2o_match |
| h2o_t |
s |
Last H2O match time |
h2o_t |
| hhmmss |
|
Time of last match |
hhmmss |
| time |
secs |
Time of last match (Seconds since Jan 1970) |
time |
MchStatus
Table 4‑27. The moment-by-moment values assocated with the current match correction, be they from a previous point match, or the on-going range match adjustments.
| MatchCO2 |
µmol mol-1 |
CO2 Match adjustment value |
MatchCO2 |
| MatchH2O |
mmol mol-1 |
H2O Match adjustment value |
MatchH2O |
| MatchValveR |
% |
Reference match valve position |
MatchValveR |
| MatchValveS |
% |
Sample match valve position |
MatchValveS |
| cf_co2_a |
mmol mol-1 |
CO2 Match coeff a |
cf_co2_a |
| cf_co2_b |
|
CO2 Match coeff b |
cf_co2_b |
| cf_co2_c |
|
CO2 Match coeff c |
cf_co2_c |
| cf_co2_d |
|
CO2 Match coeff d |
cf_co2_d |
| cf_h2o_a |
mmol mol-1 |
H2O Match coeff a |
cf_h2o_a |
| cf_h2o_b |
|
H2O Match coeff b |
cf_h2o_b |
| cf_h2o_c |
|
H2O Match coeff c |
cf_h2o_c |
| cf_h2o_d |
|
H2O Match coeff d |
cf_h2o_d |
| co2_elapsed |
min |
Elapsed time since CO2 match |
co2_elapsed |
| co2_fit_high |
µmol mol-1 |
CO2 range match fit upper limit |
co2_fit_high |
| co2_fit_low |
µmol mol-1 |
CO2 range match fit lower limit |
co2_fit_low |
| h2o_elapsed |
min |
Elapsed time since H2O match |
h2o_elapsed |
| h2o_fit_high |
mmol mol-1 |
H2O range match fit upper limit |
h2o_fit_high |
| h2o_fit_low |
mmol mol-1 |
H2O range match fit lower limit |
h2o_fit_low |
Meas
Table 4‑28. The essential measurements coming from the head that go into gas exchange computations.
| CO2_a |
µmol mol-1 |
Sample cell CO2 w/o match adjust |
CO2_a |
| CO2_r |
µmol mol-1 |
Reference cell CO2 concentration |
CO2_r |
| CO2_s |
µmol mol-1 |
Sample cell CO2 concentration |
CO2_s |
| Fan_speed |
rpm |
Chamber fan rotation rate |
Fan_speed |
| Flow |
µmol s-1 |
Flow rate into chamber |
Flow |
| H2O_a |
mmol mol-1 |
Sample cell H2O w/o match adjust |
H2O_a |
| H2O_r |
mmol mol-1 |
Reference cell H2O concentration |
H2O_r |
| H2O_s |
mmol mol-1 |
Sample cell H2O concentration |
H2O_s |
| Offset |
°C |
Offset built-in to Tleaf |
Offset |
| Offset2 |
°C |
Offset built-in to Tleaf2 |
Offset2 |
| Qamb_in |
µmol m-2 s-1 |
In-chamber ambient PPFD |
PPFD_in |
| Qamb_out |
µmol m-2 s-1 |
External ambient PPFD |
PPFD_out |
| ΔPcham |
kPa |
Chamber overpressure |
Pchamber |
| Pa |
kPa |
Atmospheric pressure |
Press |
| TIME |
s |
Seconds since Jan 1970 |
TIME |
| Tair |
°C |
Chamber air temperature |
Tchamber |
| Tleaf |
°C |
Leaf thermocouple #1 |
Tleaf |
| Tleaf2 |
°C |
Leaf thermocouple #2 |
Tleaf2 |
Meas2
Table 4‑29. Secondary head measurements, of interest, but not directly used in gas exchange computations.
| CO2_b |
µmol mol-1 |
Reference cell CO2 w/o match adjust |
CO2_b |
| ΔCO2 |
µmol mol-1 |
Sample - Reference CO2 |
CO2_d |
| CO2_r_d |
µmol mol-1 |
Reference cell CO2, dry mole fraction |
CO2_r_d |
| CO2_s_d |
µmol mol-1 |
Sample cell CO2, dry mole fraction |
CO2_s_d |
| H2O_b |
mmol mol-1 |
Reference cell H2O w/o match adjust |
H2O_b |
| ΔH2O |
mmol mol-1 |
Sample - Reference H2O |
H2O_d |
| Td_r |
°C |
Reference cell dew point |
Td_r |
| Td_s |
°C |
Sample cell dew point |
Td_s |
| e_r |
kPa |
Reference cell vapor pressure |
e_r |
| e_s |
kPa |
Sample cell vapor pressure |
e_s |
MPF
Table 4‑30. Settings and results specific to the latest MultiPhase Flash.
| ID |
|
ID of this MPF flash |
EVENT_ID |
| P1_Fmax |
|
Fmax in Phase 1 |
P1_MAXF |
| P2_dQdt |
mol m-2 s-2 |
Phase 2 dQ/dt |
P2_DQDT |
| P3_ΔF |
|
Fmeasured - Fpredicted Phase 3 |
P3_DELTAF |
| P1_dur |
ms |
Phase 1 duration |
PHASE1_DURATION |
| P2_dur |
ms |
Phase 2 duration |
PHASE2_DURATION |
| P3_dur |
ms |
Phase 3 duration |
PHASE3_DURATION |
| P1_Qmax |
µmol m-2 s-1 |
Qmax in Phase 1 |
QMAX |
pH_cal
Table 4‑31. Constants and settings associated with the Aquatic Chamber pH probe.
| Cal 1 T |
°C |
Buffer temperature |
cal1_T |
| Cal 1 Signal |
V |
Buffer signal |
cal1_V |
| Use Cal 1 |
|
|
cal1_on |
| Cal 1 pH |
pH |
Buffer pH |
cal1_pH |
| Cal 2 T |
°C |
Buffer temperature |
cal2_T |
| Cal 2 Signal |
V |
Buffer signal |
cal2_V |
| Use Cal 2 |
|
|
cal2_on |
| Cal 2 pH |
pH |
Buffer pH |
cal2_pH |
| Cal Factor |
|
Probe calibration adjustment factor |
slope |
| Cal 0 Signal |
V |
Signal at neutral pH |
std_V |
| Cal 0 pH |
|
Neutral pH |
std_pH |
QConst
Table 4‑32. Light weighting factors used in the computation of LeafQ results.
| fQ_Amb_in |
|
Fraction Q from Amb_in |
fQ_Amb_in |
| fQ_Amb_out |
|
Fraction Q from Amb_out |
fQ_Amb_out |
| fQ_ConsoleLS |
|
Fraction Q from ConsoleLS |
fQ_ConsoleLS |
| fQ_Flr |
|
Fraction Q from Flr |
fQ_Flr |
| fQ_HeadLS |
|
Fraction Q from HeadLS |
fQ_HeadLS |
Raw
Table 4‑33. Low-level readings from the head that are used in to computing the Meas and Meas2 values.
| Flow_s_v |
V |
Sample cell flow signal |
Flow_a_v |
| Flow_r_v |
V |
Reference cell flow signal |
Flow_b_v |
| VPchamber |
V |
Chamber pressure sensor signal |
VPchamber |
| Vflow |
V |
Flow meter signal |
Vflow |
| Wc_s |
|
Absorbing counts, CO2 sample |
Wc_a |
| Wc_r |
|
Absorbing counts, CO2 ref |
Wc_b |
| Wco_s |
|
Non-absorbing counts, CO2 sample |
Wco_a |
| Wco_r |
|
Non-absorbing counts, CO2 ref |
Wco_b |
| Ww_s |
|
Absorbing counts, H2O sample |
Ww_a |
| Ww_r |
|
Absorbing counts, H2O ref |
Ww_b |
| Wwo_s |
|
Non-absorbing counts, H2O sample |
Wwo_a |
| Wwo_r |
|
Non-absorbing counts, H2O ref |
Wwo_b |
| abs_c_a |
|
CO2 absorptance cell A |
abs_c_a |
| abs_c_b |
|
CO2 absorptance cell B |
abs_c_b |
| abs_h_a |
|
H2O absorptance cell A |
abs_h_a |
| abs_h_b |
|
H2O absorptance cell B |
abs_h_b |
| Tleaf2_mv |
mV |
Leaf2 temp V |
leaf2_t_mv |
| Tleaf_mv |
mV |
Leaf temp V |
leaf_t_mv |
SoilConfig
Settings used for the soil chamber.
Table 4‑34[caption]
| Dead band |
s |
Delay after closing |
DeadBand |
| # Reps |
|
Repetitions per measurement |
NumReps |
| Soil Type |
|
For Stevens Probe |
soiltype |
| Pump Speed |
% |
Circulation pump speed |
speed |
| ID Tag |
|
Measurement ID |
tag |
| Extra |
s |
Time between repetitions |
textra |
| Rep duration |
s |
Time chamber is closed |
tmax |
| Tsoil |
|
Sensor for TSoil |
tss |
SoilConst
Table 4‑35. Computational constants used for the soil chamber.
| Vcham |
cm3 |
Chamber volume |
ChamberVolume |
| Offset |
cm |
Collar height |
CollarHt |
| Ctarget |
µmol mol-1 |
Compute flux here |
Ctarget |
| Area |
cm2 |
Exposed soil area |
SoilArea |
| Virga |
cm3 |
Non-chamber Volume |
SysVolume |
| Vtotal |
cm3 |
System volume + collar volume |
TotalVolume |
SoilData
Table 4‑36. Real-time data available during a soil chamber observation.
| CO2 |
µmol mol-1 |
CO2 in soil chamber |
CO2 |
| CStatus |
|
Chamber location status |
CStatus |
| Cdry |
µmol mol-1 |
Dry CO2 in soil chamber |
Cdry |
| Elapsed |
s |
Time chamber has been closed |
Elapsed |
| Flux_to |
µmol m-2 s-1 |
Estimate of Flux_to during a measurement |
F_avg |
| H2O |
mmol mol-1 |
H2O in soil chamber |
H2O |
| MStatus |
|
Measurement status |
MStatus |
| MStatusN |
|
Measurement status index (0-5) |
MStatusN |
| Msoil |
m3 m-3 |
Soil moisture wfv |
Msoil |
| Pressure |
kPa |
Atmospheric Pressure |
Pressure |
| Progress |
% |
Percent completed of measurement |
Progress |
| RH |
% |
Relative Humidity in soil chamber |
RH |
| Rep |
|
Rep Status |
Rep |
| RepProg |
% |
Percent completed of current rep |
RepProg |
| STATE |
|
Chamber state value |
STATE |
| TIME |
s |
Time_t of data record |
TIME |
| Tair_sc |
°C |
Air temp in soil chamber |
Tchamber |
| Tsoil |
°C |
Soil temperature |
Tsoil |
| Vbellows |
V |
Bellows signal |
VBELLOWS |
| dCdry/dt |
µmol mol-1 s-1 |
Rate of change Cdry |
dCdry/dt |
SoilComp2
Table 4‑37. Computations available following a soil chamber observation.
| Cdry_o |
µmol mol-1 |
Initial Cdry |
Cdry_o |
| F_CV |
% |
CV of flux |
F_CV |
| Flux_to |
µmol m-2 s-1 |
Flux computed at Cdry_o |
F_o |
| Flux_target |
µmol m-2 s-1 |
Flux computed at target concentration |
F_target |
| MeasNum |
|
Measurement number |
MeasNum |
| Ms_avg |
m3 m-3 |
Soil moisture wfv average |
Ms_avg |
| N |
|
Number points in regression |
N |
| P_to |
kPa |
Initial atmospheric pressure |
P_o |
| RepNum |
|
Rep number |
RepNum |
| T_to |
°C |
Initial air temperature |
T_o |
| Ts_avg |
°C |
Soil temperature average |
Ts_avg |
| W_to |
mmol mol-1 |
Initial water concentration |
W_o |
| intercept |
µmol mol-1 s-1 |
Slope of dCdry/dt as function of Cdry |
intercept |
| r2 |
|
r2 dCdry/dt as function of Cdry |
r2 |
| slope |
s-1 |
Slope of dCdry/dt as function of Cdry |
slope |
Status
Table 4‑38. Primary status information on the state of the system.
| CO2_% |
% |
CO2 injector amount on |
CO2_percent |
| CO2_r_sp |
µmol mol-1 |
CO2 mixer algorithmic setpoint |
CO2_r_setpoint |
| DIAG |
|
Head diagnostic value. Normal = 2 or 3 |
DIAG |
| Desiccant_% |
% |
Desiccant valve position |
Desiccant_percent |
| Flow_r |
µmol s-1 |
Ref cell flow |
Flow_r |
| Flow_s |
µmol s-1 |
Sample cell flow |
Flow_s |
| H2O_r_sp |
mmol mol-1 |
H2O algorithmic setpoint |
H2O_r_setpoint |
| Txchg_sp |
°C |
Heat exchange algorithmic setpoint |
Heatx_setpoint |
| Humidifier_% |
% |
Humidifier valve position |
Humidifier_percent |
| SS_s |
% |
Sample cell signal strength |
SS_a |
| SS_r |
% |
Reference cell signal strength |
SS_b |
| Ts |
°C |
A cell inlet temperature |
Ta |
| Tr |
°C |
B cell inlet temperature |
Tb |
| Tchopper |
°C |
IRGA chopper temperature |
Tchopper |
| Tirga |
°C |
IRGA block temperature |
Tirga_block |
| Txchg |
°C |
Heat exchanger temperature |
Txchg |
Status2
Table 4‑39. Secondary status information on the state of the system.
| AccCO2_soda |
mg |
Absorbed CO2 (mg) by soda lime |
AccCO2_soda |
| AccH2O_des |
mg |
Absorbed H2O (mg) by desiccant |
AccH2O_des |
| AccH2O_hum |
mg |
Added H2O (mg) from humidifier |
AccH2O_hum |
| CO2_hrs |
hrs |
CO2 cartridge hours |
CO2_hrs |
| Console_H2O |
mmol mol-1 |
H2O after humidifier |
Console_H2O |
| Console_RH |
% |
RH after humidifier |
Console_RH |
| Console_T |
°C |
T after humidifier |
Console_RH_T |
| Fan_% |
% |
Fan speed in percent |
Fan_percent |
| Flow_% |
% |
Flow split position, -100 to +100 |
Flow_percent |
| Pump |
|
Pump status: 0=off, 0.5=min, 1=low, 2=med, 3=high |
Pump |
| Tleaf_j |
°C |
Tleaf junction temp |
TleafJunction |
| Tleaf2_j |
°C |
Tleaf2 junction temp |
TleafJunction2 |
| Tchp_pwm |
V |
Chopper temp PWM in V |
chopper_t_pwm_v |
| diag_12v |
V |
12v V test point |
diag_12v |
| diag_20v |
V |
20v V test point |
diag_20v |
| diag_3_3vf |
V |
3.3vf V test point |
diag_3_3vf |
| diag_5_4v |
V |
5.4v V test point |
diag_5_4v |
| diag_5va |
V |
5va V test point |
diag_5va |
| Txchg_pwm |
V |
Heat exchanger temp PWM in V |
heatx_t_pwm_v |
Stability
Table 4‑40. Content depends on what variables are being targeted for stability checking.
| ΔCO2:MN |
µmol mol-1 |
Mean of ΔCO2 |
CO2_d.Meas2:MN |
| ΔCO2:OK |
|
Stable status ΔCO2 |
CO2_d.Meas2:OK |
| ΔCO2:SD |
µmol mol-1 |
Standard deviation of ΔCO2 |
CO2_d.Meas2:SD |
| ΔCO2:SLP |
µmol mol-1 min-1 |
Rate of change of ΔCO2 |
CO2_d.Meas2:SLP |
| F:MN |
|
Mean of F |
FLUOR.FlrLS:MN |
| F:OK |
|
Stable status F |
FLUOR.FlrLS:OK |
| F:SD |
|
Standard deviation of F |
FLUOR.FlrLS:SD |
| F:SLP |
min-1 |
Rate of change of F |
FLUOR.FlrLS:SLP |
| ΔH2O:MN |
mmol mol-1 |
Mean of ΔH2O |
H2O_d.Meas2:MN |
| ΔH2O:OK |
|
Stable status ΔH2O |
H2O_d.Meas2:OK |
| ΔH2O:SD |
mmol mol-1 |
Standard deviation of ΔH2O |
H2O_d.Meas2:SD |
| ΔH2O:SLP |
mmol mol-1 min-1 |
Rate of change of ΔH2O |
H2O_d.Meas2:SLP |
| Stable |
|
Number of stable items |
Stable |
| State |
|
stable # / total # |
State |
| Total |
|
Number of items tracked |
Total |
SysObs
Table 4‑41. System values associated with the latest logged observation.
| averaging |
s |
Additional averaging on latest observation |
averaging |
| date |
|
Date & time of last observation |
date |
| elapsed |
s |
Seconds since first observation |
elapsed |
| hhmmss |
|
HH:MM:SS of last observation |
hhmmss |
| obs |
|
Number observations logged |
obs |
| time |
s |
Seconds since Jan 1970 |
time |
SysConst
Table 4‑42. System constants associated with the head computations.
| AvgTime |
sec |
Averaging time for head measurements |
AvgTime |
| Oxygen |
% |
Oxygen concentration |
Oxygen |
UserDefCon
Table 4‑43. Content depends on what user defined constants are presently active.
| myConstant |
km |
Distance from the sun |
myConstant |
UserDefVar
Table 4‑44. Content depends on what user defined variables are presently active.
| myVariable |
|
flow times area |
myVariable |
UserCal
Table 4‑45. A selection of the current head user calibration values.
| CO2_s_span1 |
|
User span 1 for sample CO2 |
co2aspan1 |
| CO2_s_span2 |
|
User span 2 for sample CO2 |
co2aspan2 |
| CO2_s_zero |
|
User zero for sample CO2 |
co2azero |
| CO2_r_span1 |
|
User span 1 for sample CO2 |
co2bspan1 |
| CO2_r_span2 |
|
User span 2 for sample CO2 |
co2bspan2 |
| CO2_r_zero |
|
User zero for reference CO2 |
co2bzero |
| Fs_zero |
|
User zero for sample flow meter |
flowazero |
| Fr_zero |
|
User zero for reference flow meter |
flowbzero |
| F_zero |
|
User zero for main flow meter |
flowmeterzero |
| H2O_s_span1 |
|
User span 1 for sample H2O |
h2oaspan1 |
| H2O_s_span2 |
|
User span 2 for sample H2O |
h2oaspan2 |
| H2O_s_zero |
|
User zero for sample H2O |
h2oazero |
| H2O_r_span1 |
|
User span 1 for sample H2O |
h2obspan1 |
| H2O_r_span2 |
|
User span 2 for sample H2O |
h2obspan2 |
| H2O_r_zero |
|
User zero for reference H2O |
h2obzero |
Note: In a data file, a value of sys::time and GasEx::Time will be different if the data was logged with extra averaging time. The time for sys::time is always the actual time data was logged, while GasEx::time lags behind by an amount that is equal to T/2, where T is the averaging time.