Measuring leaf-level gas exchange

Gas exchange theory and equation summary

In this section, we describe how the LI-6800 probes the gas exchange processes in a leaf.

Leaf-level interactions with the atmosphere

Leaves are bounded by epidermal cell layers that help to separate the leaf intercellular air space from the atmosphere. Between the epidermal layers are various other cell types, e.g., palisade, mesophyll, bundle sheath, etc., depending upon the many different varieties of leaf morphologies found in nature. As a starting point, we'll refer to a simplified schematic of a leaf cross-section throughout this preface (Figure 9‑1).

A typical plant cell has a ligneous wall and a plasma membrane that encapsulate the aqueous, cytosolic compartment. While there are many important cellular organelles, the cell depicted here is chloroplast-centric: the only organelle shown within the cell is a double membrane chloroplast, which encloses the aqueous stromal compartment. Localized within the stroma is a complex network of thylakoid membranes that also have been significantly simplified in the schematic. Integrated within the thylakoid membrane are all the proteins, pigments and redox cofactors that function to transduce light energy into the transfer of electrons and protons, the combined effect of which results in the storage of light energy in chemical forms of energy (i.e., NADPH and ATP) (6). These chemical intermediates can be used to power, among other processes, the enzymatic reduction of CO2 to the level of sugar phosphates (7).

A model plant cell showing the major components and pathways for biochemical action.
Figure 9‑1. Cross section of a simplified leaf composed of a single cell with a single organelle: the chloroplast. This leaf schematic has many of the biophysical and biochemical photosynthetic processes represented. Resistance symbols () indicate resistance to diffusion. The relative concentrations of gases are indicated by the size of characters.

Atmospheric gases (CO2, O2, H2O and others) diffuse into and out of the intercellular leaf air space through stomatal apertures (Figure 9‑2). Stomatal apertures cause resistance to gas diffusion (8). The intercellular air space is assumed to be saturated for water vapor at leaf temperature (9). The relative water vapor concentration within the leaf (Wi) is typically greater than the water vapor concentration in the atmosphere (Wa) as a result of this stomatal resistance. Stomatal resistance also results in a lower CO2 concentration in the leaf intercellular air spaces (Ci) than in the atmosphere (Ca).

Figure 9‑2. CO2 and H2O vapor move into and out of the leaf through stomatal apertures.

Leaves principally absorb light within the visible region (mostly red and blue) of the electromagnetic spectrum (10). Light absorption, which powers photosynthesis (Figure 9‑3), occurs within specialized pigment-binding protein complexes referred to as photosystems I and II (PSI and PSII) (11). These photosystems are localized within the chloroplast thylakoid membranes (12) and they function in series to facilitate linear electron transfer (ETR; µmol electrons m-2 s-1). Linear ETR is a proton-coupled process (13, 14) that establishes a trans-thylakoid proton motive force (pmf) consisting of both electric field (ΔΨ) (15) and pH (ΔpH) gradients (16).

Figure 9‑3. Photosystems I (right) and II (left), are pigment binding protein complexes where the leaf absorbs light.

Photosynthesis-driven transpiration begins with evaporation of H2O from cell wall orifices and into leaf intercellular air spaces (17). Transpiration moves H2O from below-ground soil sources to the atmosphere, transporting ions (e.g., K+ and Cl-, etc.) (18) throughout the plant. These ions serve many cellular purposes including regulation of the relative partitioning of the pmf into ΔΨ and ΔpH components (16, 19).

Energy producing and consuming reactions

The energy producing (i.e., NADPH and ATP formation) and consuming (CO2 assimilation, etc.) reactions of photosynthesis are tightly coordinated (20). The tight coordination is a function of the rigid stoichiometric relationships of key steps within the series of photosynthetic reactions. Assuming 100% quantum conversion at PSI and PSII (21), absorption of four photons at each photosystem results in the net oxidation of two H2O molecules, thereby releasing four protons (H+) into the thylakoid lumen and the generation one molecule of O2 (Figure 9‑1). The electrons freed during these thermodynamically favorable reactions can be subsequently used to form two molecules of NADPH. This transfer of electrons is a somewhat more explicit description of linear ETR.

If the Q-cycle at the cytochrome b6f complex operates constitutively (14, 22 - 24), oxidation of two H2O molecules would result in the net accumulation of twelve protons in the thylakoid lumen, thereby contributing to the pmf (13). The resulting amount of ATP generated as these twelve protons move down their electrochemical gradient through the chloroplast ATP synthase is determined by the enzyme’s H+/ATP ratio (25). The ATP synthase imposes a resistance to proton efflux from the lumen to the stroma of the chloroplast (Figure 9‑4) (6). The H+/ATP ratio of the ATP synthase is contested (26), but if the ratio is four, movement of twelve protons through the ATP synthase would result in the formation of three ATP molecules. Thus, oxidation of two H2O molecules would result in two NADPH and three ATP molecules, precisely the chemical stoichiometry needed to assimilate a single CO2 molecule (7).

Figure 9‑4. Protons encounter resistance as they move from the lumen to the stroma of the chloroplast.

In contrast, given the putative c-subunit composition of the chloroplast ATP synthase (27) (Figure 9‑5), an H+/ATP ratio of 4.7 has been suggested. The net result is a deficit in the amount of ATP produced by linear ETR alone in comparison to that needed to balance the ATP:NADPH ratio for CO2 assimilation alone. Cyclic electron transfer (ETR) around PSI has been suggested as a mechanism for augmenting the pmf, thereby accounting for the deficit of ATP produced by linear ETR (28, 29). The rate of cyclic ETR around PSI needed to accommodate for the short fall in ATP formation has been suggested to be about 12% of linear ETR (26).

Figure 9‑5. The ratio of H+ to ATP is typically 4.7.

In either case, the relationship between linear ETR and the rate of CO2 assimilation is predicted to have a ratio of slightly more than four under conditions when CO2 assimilation is the predominant sink for ATP and NADPH (30). Such conditions occur when photorespiration is inhibited by performing experiments on C4 plants or, under low O2, on C3 plants.

Photosynthetic light reactions must be highly adaptive to a dynamic environment to coordinate the energetic demands for ATP and NADPH while simultaneously regulating light capture (31). Processes other than CO2 assimilation (such as photorespiration, nitrogen and sulfur metabolism) have different ATP and NADPH requirements. Non-photochemical quenching (NPQ) of excess absorbed light energy serves as a photo-protective mechanism (32, 33, 34), preventing formation of highly reactive forms of oxygen: singlet oxygen (1O2) and the superoxide radical anion (O2-).

Energy-dependent quenching (qE) (35), the predominant component of NPQ, is a dual function of the ΔpH component of the pmf (32). The ΔpH both 1) activates a thylakoid lumen-localized enzyme violaxanthin de-epoxidase, which converts violaxanthin (V) to zeaxanthin (Z) in the PSII antenna (36); and 2) regulates protonation of thylakoid lumen-exposed, amino acid residues associated with PSII proteins (37, 38).

Dynamic coordination of the light reactions is required to meet the energetic and photoprotective demands during fluctuating biochemistries that occur during environmental stress, such as drought, for example (6, 31). The coordination is mediated by changes in cyclic ETR around PSI and changes in the proton conductance of the chloroplast ATP synthase (gH+) (19, 39). The former is thought to be predominantly involved in regulating the output ratio of ATP:NADPH, while the latter is thought to be primarily responsible for modulating qE sensitivity (6, 31).

Energy production and photoprotection

Under fluctuating biochemical demands, qE sensitivity refers to a relative increase in the amount of qE at a given flux of linear ETR (19, 40). Linear ETR and qE increase in hyperbolic and sigmoid manners, respectively, with increasing light intensity under ambient CO2 and 21% O2 (Figure 9‑6, black squares), as well as under 50 ppm CO2 and 21% O2 (Figure 9‑6, red circles). The lower CO2 concentration artificially imposes a change in biochemical demand that mimics various environmental stresses. At the highest light intensity, lowering the CO2 concentration decreased linear ETR by 42% in comparison to values under ambient CO2, whereas qE increased by approximately 2-fold under the same low CO2 concentration treatment. In essence, qE became more sensitive to linear ETR in response to the diminished biochemical demand. Combining analyses of electron and proton transfer reactions of photosynthesis has previously shown that changes in qE sensitivity are predominantly attributable to proportional changes in gH+ (19, 40).

Figure 9‑6. Linear ETR and qE both increase with increasing light intensity under ambient CO2 and 21% oxygen (squares), and under 50 ppm CO2 and 21% oxygen (circles).

Coordination of assimilation and transpiration

Optimal assimilation of CO2 from the atmosphere by leaf-level photosynthesis is directly coupled with H2O loss. Energy use from light capture, carbon assimilation, and H2O loss is coordinated through regulation of stomatal conductance (gs) (8) and mesophyll conductance (gm) (41 - 45). Research into understanding the inter-dependencies of these regulatory mechanisms has led to interest in selecting for plants to increase the gm:gs ratio, the result of which could lead to more water use efficient plants (46).

Assessing the gas exchange parameters

The gas exchange model described below assumes that the system has no leaks and that air is perfectly mixed in the chamber. In practice, this does not always happen. We present a detailed description of the computations when these assumptions are not met in Mass balance derivation using a two-compartment model.

An open gas exchange system

The LI-6800 is an open gas exchange system (9) that measures many parameters needed to estimate photosynthetic gas exchange. The gas exchange system measures the flow rate (µmol air s-1) entering the leaf chamber (µo) and the CO2 and H2O concentrations entering (c0 and w0, respectively) and exiting (ca and wa, respectively) the leaf chamber (Figure 9‑7).

Figure 9‑7. Simplified schematic of an open gas exchange system for leaf-level photosynthesis measurements.
Figure 9‑8. Flow schematic showing an overhead view of the head and leaf chamber (top), and a side view of the upper and lower cuvettes (bottom). In leaf chambers (6800-12/A and 6800-13), a photodiode measures light near the leaf surface while one or more leaf temperature thermocouples measures leaf temperature.

The LI-6800 computes the mass flow per time (i.e., µmol CO2 s-1 and mmol H2O s-1) of these gases into and out of the chamber. The differences between the CO2 and H2O concentrations into and out of the leaf chamber are due to leaf-level CO2 assimilation and transpiration on a leaf area (s) basis (µmol CO2 m-2 s-1 and mmol H2O m-2 s-1). During photosynthetic carbon assimilation (A), the leaf takes up CO2 from the air entering the chamber while simultaneously releasing H2O via transpiration (E) into the chamber air. While the flow rate of air coming out of the chamber (µa) is not directly measured, it is assumed to be:

9‑1

where s represents leaf area. Transpiration from the leaf is adding water vapor to the chamber, which changes the air density and therefore is added to the total flow into the leaf chamber.

The open-system measurements can subsequently be used to estimate important gas exchange parameters (9, 47). Transpiration, on a leaf area (s) basis, is calculated as:

9‑2

We can rearrange equations 9‑1 and 9‑2 to give:

9‑3

Similarly, we compute net CO2 assimilation (ANet) (47, 48) as:

9‑4

Using equation 9‑1, equation 9‑4 rearranges to:

9‑5

The cascade of derived parameters

In accordance with reasonable assumptions (9), we use E to derive an expression that represents the total conductance of a leaf to water vapor flux (gTw):

9‑6

In general, transpiration comes from stomata located on both sides of the leaf. On each side, water vapor diffuses through the stomata and then through the boundary layer to the mixed atmosphere. For each side of the leaf, total resistance to water vapor diffusion (rw) is the sum of stomatal resistance (rsw) plus boundary layer resistance (rbw). So,

9‑7

and

9‑8

for lower (l) and upper (u) leaf surfaces, respectively. We assume boundary layer resistance is the same on both sides of the leaf, so rbw = = . We define a stomatal ratio:

9‑9

Total stomatal resistance is a parallel combination of stomatal resistances from the upper and lower surfaces, so

9‑10

Total conductance measured by equation 9‑6 represents a parallel combination of series resistances from the upper leaf surface and lower leaf surface, so

9‑11 ,

since gi = 1/ri. From these relationships it can be shown that

9‑12 .

Or, in terms of conductances,

9‑13 .

Solving for gsw yields:

9‑14 .

Similarly, for conductance to CO2,

9‑15

where 1.6 is the ratio of the diffusivities of CO2 to water vapor in air and 1.37 is the same ratio in the boundary layer.

Lastly, we give an expression for Ci (9):

9‑16

It should be noted that equation 9‑16 is derived from the compilation of many fundamental parameters measured by the LI-6800 gas exchange system.

The Farquahar-von Caemmerer-Berry model

Development of the Farquhar-von Caemmerer-Berry (FvCB) model of photosynthesis (47) was made possible with the types of above-mentioned measured and derived parameters. The general form of the FvCB model is (48, 51):

9‑17

Equation 9‑17 expresses ANet as a function of the minimum rate of photosynthesis that can be supported by: 1) RuBisCO activity (wc); 2) energetic chemical-intermediate (i.e., ATP and NADPH) production for ribulose 1,5-bisphosphate (RuBP) regeneration (wj); and 3) triose phosphate export and utilization (wp). The CO2 compensation point in the absence of day respiration (Γ*) and day respiration (Rd) are routinely estimated from literature sources but can also be measured (52). While expressions for all three potentially limiting processes can be more fully defined (51), we will focus on the RUBP-regeneration limited expression because it directly links measurements of linear ETR and A. The full expression for RUBP-regeneration limited photosynthesis is (48, 51):

9‑18

J is the electron transport rate needed to supply energy for carboxylation and oxygenation reactions catalyzed by RuBisCO. We’ll say more about J below. The coefficients 4 and 8 can be altered depending upon the assumptions about the relative energy demands for ATP and NADPH (51).

Equation summary

The LI-6800 measures and computes the following parameters related to leaf-level gas exchange measurements.

Light

The photosynthetic photon flux density (PPFD) readings Qin and Qout are computed from

9‑19

9‑20

where Sqin and Sqout are the calibration factors, and Iqin and Iqout are the raw counts from the A/D converter.

Leaf temperature

Leaf temperatures Tl1, Tl2 (°C) for the two thermocouples are computed from signals Vl1 and Vl2 (mV) by

9‑21

9‑22

where Tj1 and Tj2 are junction temperatures (°C).

Pressure

Atmospheric pressure Pa (kPa) is reported directly from the on-board sensor.

Chamber over-pressure ΔPc (kPa) is computed from

9‑23

where Vp is the signal from the differential pressure sensor (in V, but taken to be kPa) Vpo is the calibration zero for the sensor, G is fan speed (rpm), F is flow rate, and pca and pcb are empirical coefficients that depend on chamber type (Table 9‑1).

Table 9‑1. Values for pressure correction coefficients for each chamber type.
Chamber pca pcb
6800-01 Fluorometer 0 0
6800-01A Fluorometer 4.3321E-10 1.0215E-5
6800-12 3x3 0 0
6800-12A 3x3 2.6520E-10 1.3064E-5
6800-13 6x6 4.3998E-11 1.4792E-5
6800-17 Small Plant 0 0
6800-19 Custom 0 0

Flow

The flow F (µmol s-1) to the leaf chamber is computed from

9‑24

where af1...af7 are factory calibration constants, Vf is flow meter signal (V), Vfo is calibration zero, and Tk is IRGA block temperature.

Reference and sample IRGA cell flows Fr and Fs (µmol s-1) are computed from

9‑25

9‑26

where Vfr and Vfs are cell exit flow sensor voltage, Vfro and Vfso are the offset value of that voltage, and ar1...ar4 and as1...as4 are factory calibration coefficients.

CO2

Reference and sample CO2 concentrations Cr and Cs are given by

9‑27

where Mc() is the CO2 match correction function:

9‑28

9‑29

9‑30

where αcr and αcs are absorptances, Pa is atmospheric pressure, Tr and Ts are the cell inlet temperatures, fc is a 5th order polynomial with coefficients (br1...br5 or bs1...bs5) determined at the factory, Scro; Scso and Scr1; Scs1 are the span1 and span2 user calibration settings, Wa and Wb are water concentrations, Xo is the oxygen concentration in percent. Ѱc is the band broadening function for the effect of water and oxygen on CO2, given by

9‑31

where Bwc and Boc are the band broadening coefficients for H2O and O2 respectively on CO2.

Reference and sample absorptances αcr and αcs are corrected for zero drift with temperature, span drift with temperature, and cross sensitivity with water.

9‑32

9‑33

where cr1...cr3 and cs1...cs3 are empirical coefficients for CO2 absorptance span drift determined during calibration, Icr, Icro, Ics, and Icso are the raw IRGA detector absorbing and non-absorbing readings for CO2, and Iwr, Iwro, Iws and Iwso are those for water. Xwcr and Xwcs are empirical cross sensitivity coefficients for water on CO2 determined during calibration, Zcro and Zcso are the current user CO2 zero factors, and Zcr and Zcs are the factors for CO2 zero drift with temperature, determined during calibration.

Dry mole fractions Crd and Csd are computed from

9‑34

H2O

Reference and sample H2O concentrations Hr and Hs are given by

9‑35

where Mw() is the H2O match correction function:

9‑36

9‑37

9‑38

where αwr and αws are the reference and sample absorptances, fw is a 3rd order polynomial with coefficients (dr1...dr3 and ds1...ds3) determined at the factory, Swro, Swr1 and Swso, Sws1 are the H2O span1 and span2 user calibration settings for reference and sample, Ѱo is the band broadening function for the effect of oxygen on H2O, given by

9‑39

where Bow1 and Bow2 are empirical coefficients.

Absorptances αwr and αws are corrected for zero drift with temperature, span drift with temperature, and cross sensitivity with water.

9‑40

9‑41

where wr1...wr3 and ws1...ws3 are empirical coefficients for H2O absorptance span drift determined during calibration. Xcwr, Xcws are empirical cross sensitivity coefficients for CO2 on H2O determined during calibration, Zwro and Zwso are the current user H2O zero factors, and Zwr and Zws are the factors for H2O zero drift with temperature, determined during calibration.

The vapor pressure (kPa) of the air in the reference and sample cells er and es is given by

9‑42

9‑43

Reference and sample cell dew point temperatures Tdr and Tds are given by

9‑44

9‑45

Humidity indicators

The leaf chamber vapor pressure ec (kPa) is given by

9‑46

The saturation vapor pressure esc in the leaf chamber is a function of chamber air temperature Ta:

9‑47

where es(T) is the saturation vapor pressure function:

9‑48

The relative humidity hc (%) in the leaf chamber is given by

9‑49

The vapor pressure deficit of the leaf eΔl is computed from

9‑50

where Tl is leaf temperature ().

Transpiration

The mass balance of water vapor in an open system is given by

9‑51

where s is leaf area (m2), E is transpiration rate (mol H2O m-2 s-1), ue and uo are incoming and outgoing flow rates (mol s-1) from the chamber, and we and wo are incoming and outgoing water mole factions (mol H2O (mol air)-1). Since

9‑52

we can write

9‑53

which rearranges to

9‑54

The relationships between the terms in equations 9‑51 through 9‑54 and what the LI-6800 measures are

9‑55

where F is air flow rate (µmol s-1), Ws and Wr are sample and reference water mole fractions (mmol H2O (mol air)-1), and S is leaf area (cm2). The equation that the LI-6850 uses for transpiration is thus

9‑56

Stomatal conductance

The total conductance gtw of the leaf to water vapor is

9‑57

where Wl is the molar concentration of the water vapor within the leaf (mmol H2O (mol air)-1), which is computed from the leaf temperature Tl (°C) and the total atmospheric pressure in the leaf chamber

9‑58

We assume that the total resistance for the upper ru or lower rl surface of a leaf is the sum of the stomatal resistance rs and boundary layer resistance rb of that surface

9‑59

9‑60

and that the upper and lower boundary layer resistances are the same

9‑61

and we define K to be the ratio of stomatal resistances of the two sides

9‑62

Leaf stomatal resistance rs is given by

9‑63

9‑64

9‑65

9‑66

Total conductance g (the inverse of the total resistance r) can thus be written

9‑67

9‑68

9‑69

For water vapor, the total conductance gtw is related to stomatal conductance gsw and one sided boundary layer conductance gbw by

9‑70

Solving equation 9‑68 for gsw yields

9‑71

Note that although we defined K to be a particular ratio of upper to lower stomata resistances (equation 9‑62), since K appears as , we get the same mathematical result if we had defined it the other way. In other words, K is equivalent to 1/K. Note also that the LI-6400 does not use equation 9‑71, but rather a simplified approximation:

9‑72

Boundary layer

The one sided boundary layer conductance to water vapor gbw for a broadleaf is a function of fan speed G (rpm) and leaf area S (cm2).

9‑73

where f is

9‑74

and s is forced to be

9‑75

The empirical coefficients co...c4, reference pressure Po, and leaf area limits Smin and Smax depend on chamber type.

Table 9‑2. Values for boundary layer conductance for chamber types.
Chamber co c1 c2 c3 c4 Po Smin Smax
6800-01 Flr 0.250 0.35860 -4.01816E-3 0.00451074 -0.0044762 96.9 1 6
6800-01A 6 cm2 0.578 0.5229739 3.740252E-3 -6.197961E-2 -5.608586E-3 96.9 1 6
6800-01A 2 cm2 0.572 0.3872742 -1.870584E-2 0.0 -7.37389E-3 96.9 1 2
6800-12 3x3 0.500 0.44869569 1.9000035E-3 -4.26088781E-2 -3.456516E-3 96.7 2 9
6800-12A 9 cm2 0.579 0.3210639 -1.109987E-3 5.106816E-3 -3.283688E-3 96.7 2 9
6800-12A 6 cm2FB 0.345 0.552336 -4.7985e-3 0.0 -7.3557e-3 96.7 1 6
6800-12A 6 cm2SS 0.418 0.5145466 -2.5106E-3 0.0 -8.1206E-3 96.7 1 6
6800-12A 3 cm2FB 0.188 0.5795409 -1.15295E-2 0.0 -9.7259E-3 96.7 1 3
6800-12A 3 cm2SS 0.141 0.5263354 -1.27376E-2 0.0 -1.10157E-2 96.7 1 3
6800-13 6x6 0.430 0.267827 -1.164018E-4 2.248202E-3 -5.109462E-3 96.8 6 36

Net assimilation

The mass balance of CO2 in an open system is given by

9‑76

where a is assimilation rate (mol CO2 m-1 s-1), ce and co are entering and outgoing mole fractions (mol CO2 (mol air)-1). Using equation 9‑52, we can write

9‑77

which rearranges to

9‑78

To write equation 9‑78 in terms of what the LI-6800 measures, we use equations 9‑55 and

9‑79

where Cr and Cs are sample and reference CO2 concentrations (µmol mol-1), and A is the net assimilation by the leaf (µmol m-2 s-1). Substitution yields

9‑80

9‑81

9‑82

Intercellular CO2

The intercellular CO2 concentration Ci (µmol CO2 (mol air)-1) is given by

9‑83

where gtc is the total conductance to CO2. From equation 9‑69, we can write

9‑84

where 1.6 is the ratio of the diffusivities of CO2 and water in air, and 1.37 is the same ratio for the boundary layer. This is another departure from the LI-6400, which does not use equation 9‑84, but a simplified approximation

9‑85

Energy balance

The LI-6800 provides several potential sources for leaf temperature: it can be directly measured with some combination of the two leaf thermocouples (Tl1 and Tl2), or computed indirectly from a leaf energy balance (Teb). The user can specify what combination to use via three weighting factors fT1, fT2, and fTeb:

9‑86

The energy balance temperature Teb assumes that the energy balance of a leaf in the chamber has three components: net radiation R (W m-2), sensible heat flux H (W m-2), and latent heat flux L (W m-2), and that they all sum to zero:

9‑87

We consider two components of net radiation: short wave (visible and near IR) and thermal.

9‑88

where Rabs is absorbed short wave, and Rnt is net thermal. The absorbed short wave radiation is computed by

9‑89

where Qabs is the absorbed irradiance by the leaf (µmol m-2 s-1), and k is the conversion factor for transforming (µmol m-2 s-1) to (W m-2), based on the spectral characteristics of the light source.

The net thermal balance is based on the leaf temperature Tl and the surrounding chamber wall temperature Tw, so the total radiation balance R can be written as

9‑90

where ϵ is the thermal emissivity of the leaf (usually assumed to be 0.95), and σ is the Stefan-Boltzmann constant (5.67 × 10-8 W m-2 K-1). The 2 in equation 9‑90 accounts for both sides of the leaf. Wall temperature Tw is not measured, but depends on a user-specified offset ΔTw from chamber air temperature.

9‑91

The latent heat flux L is the transpiration rate E converted to W m-2.

9‑92

The sensible heat flux H is a function of the leaf - chamber air temperature difference Tl−Ta, the specific heat capacity of the air cp (29.3 J mol-1 K-1) and the one sided boundary layer conductance for heat transfer of the leaf gbh, which is 0.92 times the boundary layer conductance for water vapor gbw.

9‑93

9‑94

Equation 9‑87 becomes:

9‑95

If we let ΔT = Tl − Ta, and note for small ΔT

9‑96

Substituting equation 9‑96 into 9‑95 and solving for ΔT yields

9‑97

The energy balance leaf temperature Teb is then

9‑98

Sensor head calibration coefficients

Table 9‑3. Sensor head calibration coefficients, with XML location.
Symbol Description XML Locator (/licor/li6850/...)
af1...af7
ar1...ar4

as1...as4
Main flow sensor
Ref flow sensor
Sample flow sensor
../factory/flowmeter/a...g
../factory/irga_a/flow/a1...a4
../factory/irga_b/flow/a1...a4
Bwc
Boc

Bow1
,
Bow2
Band broadening coefficient for water on CO2
Band broadening coefficient for oxygen on CO2
Band broadening correction for oxygen on water
../factory/bb/ch
../factory/bb/cx
../factory/bb/hx0, hx1
br1...br5
bs1...bs5
Reference CO2 calibration coefficients
Sample CO2 calibration coefficients
../factory/irga_b/co2/a1...a5
../factory/irga_a/co2/a1...a5
dr1...dr3
ds1...ds3
Reference H2O calibration coefficients
Sample H2O calibration coefficients
../factory/irga_b/h2o/a1...a3
../factory/irga_b/h2o/a1...a3
mco...mc3
mwo...mw3
CO2 match coefficients
H2O match coefficients
../cfg/match/co2_adj ... co2 adj 3
../cfg/match/h2o_adj ... h2o adj 3
pca
pcb
Pressure correction or fan speed and flow rate
Pressure correction or fan speed and flow rate
../factory/chamber/pca
../factory/chamber/pcb
σcr1...σcr3
σcs1...σcs3
σwr1...σwr3
σws1...σws3
CO2 reference absorptance span drift with temp
CO2 sample absorptance span drift with temp
H2O reference absorptance span drift with temp
H2O sample absorptance span drift with temp
../factory/irga_b/co2/s1...s3
../factory/irga_a/co2/s1...s3
../factory/irga_b/h2o/s1...s3
../factory/irga_a/h2o/s1...s3
Scro
Scr1
Scso
Scs1
Span1 for reference CO2
Span2 for reference CO2
Span1 for sample CO2
Span2 for sample CO2
../cal/co2bspan1
../cal/co2bspan2
../cal/co2aspan1
../cal/co2aspan2
Sqin
Sqout
In-chamber light sensor cal
External quantum sensor cal
../cfg/ppfdin/mult
../cfg/ppfdout/mult
Swro
Swr1
Swso
Sws1
Span1 for reference H2O
Span2 for reference H2O
Span1 for sample H2O
Span2 for sample H2O
../cal/h2obspan1
../cal/h2obspan2
../cal/h2oaspan1
../cal/h2oaspan2
Vfo
Vfro
Vfso
Zero offset for main flow meter
Zero offset for reference flow meter
Zero offset for sample flow meter
../cal/flowmeterzero
../cal/flowbzero
../cal/flowazero
Vpo Zero parameter for differential pressure sensor ../cal/chamberpressurezero
Xo Oxygen concentration, percent ../factory/cal/oxygen
Xcwr
Xcws
Xwcr
Xwcs
Cross sensitivity, CO2 on H2O, reference cell
Cross sensitivity, CO2 on H2O, sample cell
Cross sensitivity, H2O on CO2, reference cell
Cross sensitivity, H2O on CO2, sample cell
../factory/irga_b/xhc
../factory/irga_b/xch
../factory/irga_a/xhc
../factory/irga_a/xch
Zcr
Zcro
Zcs
Zcso
Zero drift with temperature for reference CO2
Zero offset for reference CO2
Zero drift with temperature for sample CO2
Zero offset for sample CO2
../factory/irga_b/z
../cal/co2bzero
../factory/irga_a/z
../cal/co2azero
Zwr
Zwro
Zws
Zwso
Zero drift with temperature for reference H2O
Zero offset for reference H2O
Zero drift with temperature for sample H2O
Zero offset for sample H2O
../factory/irga_b/z
../cal/h2obzero
../factory/irga_a/z
../cal/h2obzero

Sensor measurements and computations

Table 9‑4. Sensor measurements and computations, and where to find them in the Data Dictionary.
Symbol Description (units) Name (Label) Group
αcr
αcs
αwr
αws
Reference cell CO2 absorptance
Sample cell CO2 absorptance
Reference cell H2O absorptance
Sample cell H2O absorptance
abs_c_b
abs_c_a
abs_h_b
abs_h_a
Raw
Raw
Raw
Raw
co...c4 Boundary layer function coeffs blc_a...blc_e ChambConst
Ca
Cb
Cr
Crd
Cs
Csd
∆Pc
Sample cell CO2, not adjusted for match
Sample cell CO2 (µmol mol−1)
Reference cell CO2 (µmol mol−1)
Reference cell CO2, dry mole fraction
Sample cell CO2 (µmol mol−1)
Sample cell CO2, dry mole fraction
Chamber over pressure (kPa)
CO2_a
CO2_b
CO2_r
CO2_r_d
CO2_s
CO2_s_d
Pchamber (∆Pcham)
Meas
Meas2
Meas
Meas
Meas
Meas
Meas
er
es
Icr
Icro
Ics
Icso
Reference cell vapor pressure (kPa)
Sample cell vapor pressure (kPa)
Reference CO2 raw detector count
Reference CO2 raw detector reference count
Sample CO2, raw detector count
Sample CO2 raw detector reference count
e_r
e_s
Wc_r
Wco_r
Wc_s
Wco_s
Meas2
Meas2
Raw
Raw
Raw
Raw
Iqin
Iqout
In-chamber PPFD sensor raw counts
External quantum sensor raw counts
   
Iwr
Iwro
Iws
Iwso
Reference H2O raw detector count
Reference H2O raw detector reference count
Sample H2O raw detector count
Sample H2O raw detector reference count
Ww_r
Wwo_r
Ww_s
Wwo_s
Raw
Raw
Raw
Raw
Pa Atmospheric pressure (kPa) Press (Pa) Meas
F
Fr
Fs
Flow to chamber (µmol s−1)
Flow from reference cell (µmol s−1)
Flow from sample cell (µmol s−1)
Flow
Flow_r
Flow_s
Meas
Status
Status
G Chamber fan rotation rate (rpm) Fan speed Status
Qin
Qout
In-chamber PPFD
External PPFD
PPFD_in (Q_amb_in)
PPFD_out (Q_amb_out)
Meas
Meas
Ta Leaf chamber air temperature (C) Tchamber (Tair) Meas
Tdr
Tds
Dewpoint temperature reference cell (C)
Dewpoint temperature sample cell (C)
Td_r
Td_s
Meas2
Meas2
Tj1
Tj2
Tk
Leaf T/C 1 junction temperature (C)
Leaf T/C 2 junction temperature (C)
IRGA block temperature (C)
Tleafjunction (Tleaf_j)
Tleafjunction2 (Tleaf2_j)
Tirga_block (Tirga)
Status2
Status2
Status
Tl1
Tl2
Tr
Ts
Leaf temperature 1 (C)
Leaf temperature 2 (C)
Reference cell inlet temperature (C)
Sample cell inlet temperature (C)
Tleaf
Tleaf2
Tb (Tr)
Ta (Ts)
Meas
Meas
Status
Status
Vf
Vfr
Vfs
Signal (V) from main flow sensor
Signal (V) from reference flow sensor
Signal (V) from sample flow sensor
Flow
Flow_b_v (Flow_r_v)
Flow_a_v (Flow_s_v)
Raw
Raw
Raw
Vl1
Vl2
Leaf temperature 1 signal (mV)
Leaf temperature 2 signal (mV)
leaf_t_mv (Tleaf_mv)
leaf2_t_mv (Tleaf2_mv)
Raw
Raw
Vp Differential pressure signal (V or kPa) VPchamber Raw
Wa
Wb
Wr
Ws
Sample cell H2O not corrected for match
Reference cell H2O
Reference cell H2O (mmol mol−1)
Sample cell H2O (mmol mol−1)
H2O_a
H2O_a
H2O_r
H2O_s
Meas
Meas2
Meas
Meas