Specifications
LI-720 Carbon Flux Sensor
General
-
Weight: 2.02 kg
-
Power Requirements:
-
Voltage: 9 – 33 VDC nominally; Can be powered from a standard USB port (2.0 or newer) for calibration and firmware updates. Outdoor operation from a 5 V USB power supply is not supported.
-
Power Consumption: 1.5 W
-
-
Communication Interface: SDI-12, RS-232 (10 Hz data); compatible with LI-COR IoE Modules and some data loggers
-
Operating Temperature Range: -25 – 50 °C
CO2 measurement
-
Calibration Range: 0 to 1500 µmol mol-1
-
Accuracy: Within 1.5% of reading
-
Zero Drift per °C: ±0.15 ppm typical, ±0.3 ppm maximum
-
RMS Noise: @10 Hz: 1.0 ppm (typical @ 400 µmol mol-1 CO2)
-
Direct Sensitivity to H2O: ±2.00E-05 typical, ±4.00E-05 maximum
-
Reporting Frequency: 10 Hz (samples per second)
H2O measurement
-
Calibration Range: 0 to 60 mmol mol-1
-
Accuracy: Within 1.5% of reading
-
Zero Drift (per °C): ±0.03 mmol mol-1 typical, ±0.05 mmol mol-1 maximum
-
RMS Noise: @10 Hz: 0.05 mmol mol-1 (typical @ 10 mmol mol-1 H2O)
-
Direct Sensitivity to CO2: ±0.02 typical, ±0.05 maximum
-
Reporting Frequency: 10 Hz (samples per second)
Wind speed measurements
-
Measurement Axes: U, V, W
-
Measurement Range: 0 – 30 m s-1 (horizontal wind)
-
W Offset at Zero Wind: ±0.06 m s-1
-
W Noise (1 standard deviation): <0.1 m s-1 at 5 m s-1, <0.15 m s-1 at 15 m s-1
-
Sonic Temperature Accuracy: ±0.7 °C maximum offset at 20 °C
-
Reporting Frequency: 10 Hz (samples per second)
Biomet measurements
-
Photosynthetic Photon Flux Density:
-
Range: 0-3000 µmol m-2 s-1
-
Accuracy: ±5% of reading
-
Cosine Correction: Corrected up to 75° angle of incidence
-
-
Air Temperature:
-
Range: -40 to 60 °C
-
Accuracy: ±1.5 °C; No solar load conditions
-
-
Atmospheric Pressure:
-
Range: 50 – 110 kPa
-
Accuracy: ±0.2 kPa
-
-
Relative Humidity:
-
Range: 0-100% – non-condensing
-
Accuracy: ±1% typical
-
Position and orientation
-
GPS: Provides location and precise time, including for edge flux processing
-
Magnetometer: Reports orientation relative to magnetic north (°)
-
Accelerometer: Reports instrument tilt (°)
IoE Module
General
- Operating Temperature Range: -25 to 50 °C
- Operating Humidity Range: 0 to 95%, non-condensing
- Storage Temperature Range: -40 to 65 °C; 85% RH
- Ingress Protection: IP24 Rating
- Enclosure Flame Rating: UL94-V-0
- GNSS (GPS) Support: GPS receiver for time synchronization and location information.
- Display: 35 x 55 mm; monochrome
Auxiliary power input
- Input Voltage Range: 9 to 33 VDC (>11 V recommended for startup)
- Maximum Current: 3.0 A
Digital inputs/outputs
- SDI-12 ports: 2
- Output Voltage: 12 V
- Maximum Total Current: 2.5 A
- Sensor Ports (SDI-12 'plus'): 1 (combined SDI-12 with fast serial T/R)
- USB port: Micro-B connector
Structure and load
- Ground Area Required:
- Without Guy Wires: 42 × 107 cm (16.25 × 42 inches)
- With Guy Wires at 15°: 244 × 244 cm (96 × 96 inches)
- Maximum Load on Mast: 4.54 kg (10 lbs.)
- Maximum Battery Weight: 45.4 kg (97 lbs.)
- Soil Types for Anchors: 1 to 4
Weights
- System (total): <45.4 kg (<100 lbs.)
- Battery: 24.9 kg (55.0 lbs.)
- Solar Panel: 7.5 kg (16.5 lbs.)
- Frame: 13 kg (28.7 lbs.)
Data
- Data Storage: Removable Micro SD card; 8 GB. If the modem is offline, data are stored until connectivity is restored (storage for 2 weeks of data).
- Data Format: Text; Comma-separated values (CSV)
Cellular
- Wireless Protocols: LTE CAT1
Specifications subject to change without notice.