Initial assembly

This section covers the basic steps you will follow to assemble the LI-7500RS for the first time. You'll need the analyzer head, LI-7550, Temperature Sensor, SmartFlux Core System, and the accessories.

Preparing the enclosure

Start by assembling the enclosure hardware. Keep the main power cable disconnected from the power source or de-energize the power source until the component wiring is complete, and leave the breakers in the OFF position until you are ready to power on the system.

Installing the system power supply cable

Connect the main power supply conduit to the small opening in the enclosure. Orient the lock-nut teeth so that they make contact with the enclosure surface. Connect the main power supply wires to terminals 3, 4 and 5.

Connect the main power supply conduit to the enclosure and attach each wire to the proper terminal - red to red, green to green, and black to black..
Table 2‑1. Power wire connections for the main power supply to the enclosure.
From Wire Color To
Main power (-) black DIN Terminal 3 (bottom)
Earth ground green DIN Terminal 4 (bottom)
Main power (+) red DIN Terminal 5 (bottom)

In both the eddy covariance system enclosure (7900-050) and the Biomet Data Acquisition System enclosure (7900-126), install the three strain-relief couplings in the three bottom openings.

Attach the locknuts to the fittings

In the eddy covariance system enclosure (7900-050), install the Power Distribution Kit (7900-235) on the raised DIN rail. Connect the red wire lead between the top of the breaker and top right red terminal. Install one shorting block to connect all black terminals and another to connect all red terminals.

Power wires attach to the connections as shown.
Figure 2‑1. The power distribution kit (7900-235) installed in the enclosure (7900-050).

Preparing the SmartFlux Core System and network switch

Follow the steps below when installing the SmartFlux Core System in the 7900-050 Eddy Covariance System Enclosure or 7900-126 Biomet Data Acquisition System Enclosure. If you are using a different enclosure, connect the power and data cables similarly to what is described here.

Installing the USB drive

Data are recorded to the USB drive. Install the USB drive in the USB port on the SmartFlux Core System.

Insert a USB storage device into the SmartFlux System

Connecting the GPS antenna cable

Route the GPS cable through the left opening at the bottom of the enclosure and plug it into the GPS port on the SmartFlux Core System.

Connect the GPS antenna cableConnect the GPS antenna cable

Connecting power and data cables to the SmartFlux Core System

Connect wire leads to the positive and negative terminals on the SmartFlux Core System. The black lead connects to negative (-), the red lead connects to positive (+). Plug in the 30 cm Ethernet cable that came with the SmartFlux Core System. There are two connections for positive (+) and two for negative (-). Use only one pair of +/- terminals.

Connect the power wires

Connecting power wires to the Brainboxes network switch

Power wires connect to the positive and negative terminals on the switch. The black lead connects to terminal 1 (-), the red lead connects to terminal 2 (+).

Brainboxes power wires

Installing the SmartFlux Core System and network switch

Install the SmartFlux Core System onto the left side of the upper-right DIN rail. Insert the spring-loaded clip on the bottom edge of the DIN rail and compress the spring. Rotate it into place. Mount the switch next to it.

Mount the SmartFlux SystemInstall the network switch
Biomet enclosure assembled

Connecting component power wires

The 7900-050 and 7900-126 enclosures present different wiring options. Be sure the power supplied to the enclosure is within the upper voltage limits of attached components, but capable of providing 4 to 5 amps. Keep the breaker OFF while connecting power wires.

7900-050 Eddy Covariance System Enclosure

With the 7900-050 enclosure, keep the breaker OFF. Connect the positive (+) leads to red terminals and negative (-) leads to black terminals for all components of the system. Proceed to Connecting gas analyzer power wires.

7900-126 Biomet Data Acquisition System Enclosure

With the 7900-126 enclosure, each component is powered individually from the power supply terminals, the DC-DC converter, or the DRM power outputs. 24-volt and 12-volt power supplies will be wired differently.

Caution: Do not short circuit the power supplies. For example, avoid connecting the V4 POWER OUT terminals from a Data Retention Module to the Power In terminals on a device that is powered from another power supply. Doing so will damage components and may present risk of a fire.

Select the power supply option that describes your system:

24 VDC power supply with Data Retention Module

12 VDC power supply with Data Retention Module

12 VDC power supply with no Data Retention Module

24 VDC power supply with the TDK-Lambda DC-DC converter

Preparing the gas analyzer

Begin by installing the temperature sensor, head cable, and USB drive.

Connecting the temperature sensor

The air temperature sensor (part number 9975-026) connects to the SENSOR connector on the LI-7550 (see Figure 2‑5).

Connecting the head cable

The head cable connects between the bottom of the head and the IRGA connector on the LI-7550. Align the notches on the cable connectors, then push in and turn the connector clockwise until it is tight. Continue to push in while tightening the connector until the connector is fully seated. The gasket must be compressed to ensure a watertight seal.

Connect the temperature sensor to the SENSOR port. head cable to LI-7550.
Figure 2‑5. First install the temperature sensor. Then install the head cable. Align the notches on the cable connectors, then tighten while pressing the connectors together.

Installing the LI-7550 USB drive

Open the cover to the LI-7550 and insert the USB drive from the spares kit into the USB port. Close and latch the cover.

Caution: A USB drive must be installed in the LI-7550 to log data. Beginning with SmartFlux firmware v2.2.8, a USB drive must also be installed in the SmartFlux Core System to ensure successful data processing. A full dataset is stored on both USB drives.

Connecting gas analyzer power wires

Power wires for the gas analyzer connect to terminals in the enclosure. With the eddy covariance system enclosure (7900-050) and Power Distribution Kit (7900-235), connect the red and black wire leads to corresponding red and black terminals. With the Biomet Data Acquisition System enclosure (7900-126), connect the leads as described in Table 2‑6.

Table 2‑6. Power wire connections in the Biomet Data Acquisition System.
From Wire Color To
Terminal 17 (bottom) black LI-7550 Power (-)
Terminal 19 (bottom) red LI-7550 Power (+)

Connecting Ethernet cables

Ethernet cables connect the SmartFlux Core System and LI-7550 to the network switch. Vacant ports on the network switch are for other networked components, such as the cellular gateway, PhenoCam, LI-7700, or connecting a PC.

Figure 2‑6. Install Ethernet cables to connect the SmartFlux Core System and LI-7550 to the Network Switch.

Connecting sonic anemometer data and power cables

The SmartFlux Core System can record digital data from a number of sonic anemometer models. These are described below. Power for the sonic anemometer is provided through the SmartFlux Core System.

Gill WindMaster or WindMaster Pro

Gill HS-50 or R3-50

Campbell Scientific CSAT3

Campbell Scientific CSAT3B

See CSAT3B and CSAT3 for details on the USB connection.

Metek Cage or Class-A

RM Young 81000x