Element descriptions
In this section we break down each subset of the XML tags. We begin with the root-level tags and then describe the subordinate tags in greater detail.
| Root | Parent | Value | Definition |
|---|---|---|---|
<LI850>
|
<DATA>
|
Float | See Data tags. |
<AUXDATA>
|
Float | See Auxiliary data tags. | |
<CFG>
|
Various | See Configuration tags. | |
<PUMP>
|
Various | Pump status. See Pump tags. | |
<SOURCE>
|
Integer | Hours the source has operated. See Source tag. | |
<CAL>
|
Various | See Calibration tags. | |
<POLY>
|
Various | See Polynomial tags. | |
<RS232>
|
True | False | See RS-232 serial tags. | |
<VER>
|
String | Software version. See Software version. | |
<SERIALNUM>
|
String | Serial number. See Serial number. | |
<ACK>
|
True | False | Acknowledgment of a command. | |
<ERROR>
|
String | Error. Includes a message. | |
<TECH>
|
Various | Diagnostics. See Tech tags. |
Sending this command will return the full configuration file and measured values that are configured to be sent over the serial connection.
<LI850>?</LI850>
The instrument will reply by sending the data and the confirmation message:
<LI850><ACK>TRUE</ACK></LI850>
Data tags
The <DATA> tag contains all data values sent from the instrument. All of the elements within <DATA> are readable only. The <H2O>-related tags apply to the LI-850 only.
| Root | Parent | Children | Values | R/W | Comments |
|---|---|---|---|---|---|
<LI850> or <LI830>
|
<DATA>
|
<CO2>
|
Float | R | CO2 in ppm |
<CO2ABS>
|
Float | R | CO2 absorption | ||
<H2O>
|
Float | R | H2O in mmol/mol | ||
<H2ODEWPOINT>
|
Float | R | H2O in °C | ||
<H2OABS>
|
Float | R | H2O absorption | ||
<CELLTEMP>
|
Float | R | Cell temperature | ||
<CELLPRES>
|
Float | R | Cell pressure | ||
<IVOLT>
|
Float | R | Input voltage | ||
<FLOWRATE>
|
Integer | R | Measured flow rate | ||
<RAW>
|
See Raw data tags | ||||
To query the data, send:
<li850>
<data>?</data>
</li850>
If a data value is not returned, it is probably "turned off" in the serial data stream. See RS-232 serial tags for more information.
Raw data tags
The raw detector counts can be retrieved by querying the parent <data> tag. The <h2o>-related tags apply to the LI-850 only. To see the raw detector counts, send:
<li850>
<data>
<raw>?</raw>
</data>
</li850>
To disable the raw data from the serial data stream, you have to modify the instructions under the <rs232> tag. So, send:
<li850>
<rs232>
<raw>false</raw>
</rs232>
</li850>
Auxiliary data tags
The <AUXDATA> tag contains all auxiliary data values that can be retrieved from the instrument. All of the elements within the <AUXDATA> tag are readable only.
| Root | Parent | Children | Values | R/W |
|---|---|---|---|---|
<li850>
|
<auxdata>
|
<pca>
|
Float | R |
<pha>
|
Float | R | ||
<psi>
|
Float | R | ||
<bb_eff>
|
Float | R |
To query the auxdata, send:
<li850>
<auxdata>?</auxdata>
</li850>
RS-232 serial tags
The <RS232> tag is where you set which values are output in <DATA>. TRUE means that the value will be sent; FALSE means it will not be sent. The <H2O>-related tags apply to the LI-850 only.
Sending this command will configure the output rate and the variables output over the serial connection.
<LI850>
<CFG>
<OUTRATE>1</OUTRATE>
</CFG>
<RS232>
<CO2>TRUE</CO2>
<CO2ABS>TRUE</CO2ABS>
<h2o>false</h2o>
<h2odewpoint>false</h2odewpoint>
<h2oabs>false</h2oabs>
<CELLTEMP>TRUE</CELLTEMP>
<CELLPRES>TRUE</CELLPRES>
<IVOLT>TRUE</IVOLT>
<ECHO>FALSE</ECHO>
<STRIP>FALSE</STRIP>
<flowrate>false</flowrate>
</RS232>
</LI850>
Configuration tags
Configuration tags are used to apply settings to the instrument.
| Root | Parent | Children | Values | R/W | Comments |
|---|---|---|---|---|---|
<li850>
|
<cfg>
|
<outrate>
|
0.5 to 20 | R/W | Output data every N seconds (0.5 second increments). |
<heater>
|
True | False | R/W | Turn heater on/off. | ||
<pcomp>
|
True | False | R/W | Pressure compensation on/off. | ||
<filter>
|
0 to 20 | R/W | Set a 0 to 20 second filter for data logging. | ||
<bench>
|
14 | R | Bench length (not used). | ||
<alarms>
|
Alarm settings. See Alarm tags. | ||||
<dacs>
|
DAC output settings. See DAC tags. | ||||
Sending this command will set the output rate to 1 per second, turn the optical bench heater on, apply the pressure compensation, and turn off data filtering:
<LI850>
<CFG>
<OUTRATE1</OUTRATE>
<HEATER>TRUE</HEATER>
<PCOMP>TRUE</PCOMP>
<FILTER>0</FILTER>
</CFG>
</LI850>
Alarm tags
Alarm tags configure the alarm states.
| Root | Parent | Parent | Children | Values | R/W | Comments |
|---|---|---|---|---|---|---|
<li850>
|
<cfg>
|
<alarms>
|
<enabled>
|
True | False | R/W | Enable alarms |
<source>
|
co2 | h2o | R/W | Channel (CO2, H2O) | |||
<high>
|
Float | R/W | High alarm on at this value | |||
<hdead>
|
Float | R/W | High alarm off at this value | |||
<low>
|
Float | R/W | Low alarm on at this value | |||
<ldead>
|
Float | R/W | Low alarm off at this value |
To configure the alarms, send:
<LI850>
<CFG>
<ALARMS>
<ENABLED>{bool}</ENABLED>
<SOURCE>H2O | CO2<SOURCE>
<HIGH>{float}</HIGH>
<HDEAD>{float}</HDEAD>
<LOW>{float}</LOW>
<LDEAD>{float}</LDEAD>
</ALARMS>
</CFG>
</LI850>
DAC tags
DAC tags configure the analog outputs.
| Root | Parent | Parent | Children | Values | R/W | Comments |
|---|---|---|---|---|---|---|
<li850>
|
<cfg>
|
<dacs>
|
<range>
|
2.5 | 5.0 | R/W | Output voltage. 0 to 2.5V or 0 to 5.0V |
<d1>
|
Various | R/W | CO2 | H2O | H2ODEWPOINT | CELLPRES | CELLTEMP | NONE from DAC1 | |||
<d1_0>
|
Float | R/W | Value where DAC1 outputs 0V | |||
<d1_F>
|
Float | R/W | Value where DAC1 outputs full scale | |||
<d2>
|
Various | R/W | CO2 | H2O | H2ODEWPOINT | CELLPRES | CELLTEMP | NONE from DAC2 | |||
<d2_0>
|
Float | R/W | Value where DAC2 outputs 0V | |||
<d2_F>
|
Float | R/W | Value where DAC2 outputs full scale |
To configure the DACs, send:
<LI850>
<CFG>
<DACS>
<RANGE>{2.5|5.0}</RANGE>
<D1>{NONE|CO2|H2O|H2ODP|CELLTEMP|CELLPRES}</D1>
<D1_0>{float}</D1_0>
<D1_F>{float}</D1_F>
<D2>{NONE|CO2|H2O|H2ODP|CELLTEMP|CELLPRES}</D2>
<D2_0>{float}</D2_0>
<D2_0>{float}</D2_F>
</DACS>
</CFG>
</LI850>
Pump tags
Pump tags are used to control the LI-COR pump.
To enable the pump, send:
<li850>
<PUMP>
<ENABLED>true</ENABLED>
</PUMP>
</li850>
To reset the pump hours (after replacing it, for example), send:
<li850>
<PUMP>
<HOURS>0</HOURS>
</PUMP>
</li850>
Source tag
The source tag indicates the number of hours that the optical source has operated. This information is useful for diagnostic purposes.
| Root | Parent | Children | Values | R/W | Comments |
|---|---|---|---|---|---|
<li850>
|
<source>
|
<time>
|
Integer | R/W | Number of hours of source operation |
To read the source time, send:
<li850>
<source>?</source>
</li850>
Calibration tags
Calibration tags are used to start a calibration or read calibration information. When calibrating, <date> must be paired with a <co2zero> or a <co2span>, or a <h2ozero> or <h2ospan>. The <h2o>-related tags apply to the LI-850 only.
Calibration instructions are given in Zero and span settings. To read the current calibration, send:
<li850>
<cal>?</cal>
</li850>
Polynomial tags
Polynomial tags contain calibration information from the instrument. The values included are determined at the factory. Do not change the polynomial values because that will adversely affect the analyzer performance. The <h2o>-related tags apply to the LI-850 only.
To read the polynomial information, send:
<li850>
<poly>?</poly>
</li850>
Tech tags
Tech tags indicate information about the instrument, including pump and optical source information. They are not included in the status information returned by the instrument. You can poll the tech information specifically.
To read the tech information, send:
<li850>
<tech>?</tech>
</li850>