Introduction to the LI-8250
Welcome, and thank you for your purchase of the LI-8250 Multiplexer. This document provides comprehensive operating instructions for the LI-8250 for soil gas flux measurements, flask measurements, and other applications. It describes soil chambers, gas analyzers, and accessories for the system, and it introduces SoilFluxPro Software, which is used to evaluate data collected by the LI-8250. If you have just taken delivery of the LI-8250, check the packaging list to be sure you received everything that you ordered.
LI-8250 Multiplexer
Part number LI-8250
The LI-8250 is the control unit for the system. It features a web interface for configuration from a Smart Phone, tablet, or computer with a web browser. Externally, it features a sealed connector panel for chambers or extension manifolds and one for gas analyzers and accessories.
The LI-8250 Multiplexer facilitates the connection of one or more gas analyzers with multiple chambers, providing the controller and data management for the system. Featuring embedded flux processing software, GPS, and internal storage, the LI-8250 Multiplexer can be used with LI-COR gas analyzers for real-time flux calculations or third-party gas analyzers for measure other gases.
By itself, the LI-8250 Multiplexer can connect with eight chambers. With one or more 8250-01 Extension Manifolds, the maximum number of chambers can be increased to 36. You configure and read data from the LI-8250 with the web browser of a Smart Phone, tablet, or computer using the LI-8250 built-in Wi-Fi or an Ethernet connection.
What's what
The LI-8250 has a main panel for gas analyzers and a port panel.
Main connection panel
The LI-8250 main connection panel features the power-in connector, three RJ-45 connectors for networked devices, such as trace gas analyzers, a USB connector for either the LI-870 or the 8250-02 Calibration Manifold, an auxiliary power-out connector for either the LI-870 or 8250-02 calibration manifold, the air in and air out for gas analyzers.
- Power In: Connects to the power cable (8150-706) or the optional outdoor AC-to-DC power supply (8250-770).
- Network Ports: Connects to trace gas analyzers, network, or cellular modem.
- USB Port: Connects to the LI-870 or 8250-02 Calibration Manifold.
- Power Out: Provides power to the LI-870 or the 8250-02 Calibration Manifold.
- Air Out: To the gas analyzer air inlet or T-splitter if using multiple analyzers.
- Air In: From the gas analyzer air outlet or T-splitter if using multiple analyzers.
Caution: The power output is 12 VDC with a center-positive pin, 2 A maximum. It powers either the LI-870 CO2/H2O Analyzer or the 8250-02 Calibration Manifold. Only use the included power cable (9982-010). Do not attempt to power any other devices with the PWR OUT connector. Drawing a current in excess of 2 A will open an internal circuit breaker, requiring a power cycle to reset.
Port connection panel
The LI-8250 has ports to connect with up to eight devices at a time, such as the 8250-01 Extension Manifold, the 8200-104 Opaque Long-Term Chamber, the 8200-104C Clear Long-Term Chamber, and user-built custom chambers.
- Ports: The designated port number for the connected device. Each port has air in, air out, and a power/data connector.
- Air In: Connects to the chamber or an extension manifold air outlet.
- Air Out: Connects to the chamber or an extension manifold air inlet.
- Power and Communication: Accepts the power and communication cable for extension manifolds or chambers.
Internal components
The internal components control the flow of air between analyzers, extension manifolds, and chambers.
- Pump: Provides a steady flow of air through the system.
- Port Valve Manifold: Controls the flow to and from each port.
- System Valve Manifold: Controls the flow to and from the valve manifold and gas analyzers.
- USB Ports: For connecting to an external Wi-Fi adapter or USB mass storage device.
- Network Activity LEDs: Illuminated when connected; blinks when sending or receiving communication.
- Ready LED: Steady light when the LI-8250 Multiplexer has finished booting up.
- Power LED: Illuminates when the LI-8250 Multiplexer is receiving power.
- Alternate Power Input: For connecting to the indoor AC-to-DC power supply (8250-772).
Included accessories and spares
The items listed are included with every LI-8250 Multiplexer.
Bare-leads power cable
Part number 8150-706
The bare-leads power cable is a 3-meter cable with bare wire leads that allows you to connect the LI-8250 Multiplexer to a user-supplied DC power source that reliably meets your power requirements. See Power requirements and consumption to estimate your total system power requirements.
Indoor AC-to-DC power supply
Part number 8250-772
The indoor AC-to-DC power supply provides a constant 24 VDC 80 W power using mains power. It is intended for indoor use only.
Ethernet cable
Part number 616-06116
This cable has two unsealed RJ-45 connectors and is used to connect the LI-8250 Multiplexer to a computer. Because the connectors are unsealed, this cable is not intended for field use. Ethernet cables may also be included in the optional cable assemblies, but these cables are sealed and strain-relieved for field use.
Spares kit
Part number 9982-072
This kit includes additional components and spare parts for the LI-8250 Multiplexer and 8250-01 Extension Manifold.
Optional accessories
The following accessories are not included with the LI-8250 Multiplexer, but that are available from LI-COR for your soil gas flux system.
Outdoor AC-to-DC power supply
Part number 8250-770
When used with the LI-8250, the outdoor AC-to-DC power supply provides a constant 12 VDC 80 W power source. It is weather-resistant, O-ring sealed, and is designed for long-term outdoor deployment.
Sealed-to-standard Ethernet cables
Part numbers: 392-19109 or 392-19110
Sealed-to-standard Ethernet cables are available in lengths of 5 meters (part number 392-19109) or 25 meters (part number 392-19110). These cables have one sealed RJ-45 connector and one unsealed RJ-45 connector. They are used to connect the LI-8250 (sealed) to your site network (standard).
Cable and tube assembly for chambers and extension manifolds
Part number 9982-056
This cable assembly includes a 15-meter power and data cable with sealed connectors (part number 392-18852) and two 15-meter lengths of Bev-A-Line® tubing with quick-connect fittings. It provides power, communication, and gas flow between the LI-8250 Multiplexer, 8250-01 Extension Manifolds, and long-term chambers.
Chamber options
Depending on the system you ordered, you may have one or more of the 8200-104 Opaque Long-Term Chamber or 8200-104C Clear Long-Term Chamber. A patented pressure vent at the top of the chamber prevents pressure spikes when the chamber closes and maintains chamber pressure at ambient levels under calm and windy conditions. Chambers feature a built-in thermistor for a chamber air temperature measurement, which is used in flux calculations. Each soil chamber includes soil collars, a spares kit, and replacement gasket kit described below.
8200-104 Long-Term Soil Chamber
Part number 8200-104
The 8200-104 Opaque Long-Term Soil Chamber has a white enamel chamber bowl that rotates to and away from the soil collar. The 8200-104 is weatherized and intended to be left outdoors for extended periods of time to make automated measurements.
8200-104C Clear Long-Term Soil Chamber
Part number 8200-104C
The 8200-104C Clear Long-Term Soil Chamber has a clear chamber bowl that rotates to and away from the soil collar. The 8200-104C is weatherized and intended to be left outdoors for extended periods of time to make automated measurements. The clear chamber allows sunlight to reach plants in the collar, making this chamber suitable for measurement of Net Ecosystem Exchange.
Soil collars
Part number 6581-044
Each 8200-104/C chamber includes two soil collars. Soil collars are a short section of pipe that provides a seal between the chamber and the soil and should be installed at the measurement site at least 6 hours prior to measurements (although 24+ hours is better). Additional collars can be purchased from LI-COR or made in a shop if you have access to suitable materials (see Making collars for LI-COR soil chambers).
8200-104/C Chamber spares kit
Part number 8100-613
Each chamber includes one chamber spares kit. This kit includes additional components and spare parts for each 8200-104/C Long-Term Chamber.
| Description | Quantity | Part Number |
|---|---|---|
| T-handle nut | 4 | 165-00140 |
| Bumper | 4 | 191-08069 |
| Cable tie | 9 | 218-08499 |
| Urethane tubing (1.8 m) | 1 | 222-00303 |
| Cap plug | 5 | 620-08297 |
8200-104/C Replacement gasket kit
Part number 8100-612
Each chamber includes one gasket kit that includes the components needed to replace the chamber gasket.
| Description | Quantity | Part Number |
|---|---|---|
| Collar gasket | 2 | 6581-060 |
| Chamber gasket trim (1.8 m) | 1 | 224-07606 |
| Loctite adhesive | 1 | 208-05786 |
| 6-32 x 3/8" screw | 4 | 122-00009 |
| 4-40 x 3/8" screw | 18 | 122-01578 |
8200-105/C Automated Canopy Chambers
Automated Canopy Chambers provide a large area and volume and are optimized for measuring plant-mediated fluxes, greenhouse gas (GHG) fluxes, Net Ecosystem Exchange (NEE) measurements, and a range of measurements that require a large volume or area. Automated Canopy Chambers are compatible with the LI-8250 Multiplexer (firmware v2.2.0 or higher required) and a variety of gas analyzers (LI-COR gas analyzers recommended). The height of 8200-105 and 8200-105C chambers can be extended up to 135 cm with base extensions. Automated Canopy Chambers are also compatible with 8250-01 Extension Manifold, 8250-02 Calibration Manifold, and in combination with LI-COR chambers and other chamber options.
Each Automated Canopy Chamber includes either a 10 cm or 20 cm tall soil collar and possibly one or more base extensions listed in Table 1‑2. Chambers are available with a transparent or opaque white dome.
Automated Canopy Chambers are described in Automated Canopy Chambers.
LI-190R Quantum Sensor and leveling stake
Part number 190R-8200
The LI-190R Quantum Sensor can be connected to a chamber to measure photosynthetically active radiation at the chamber to supplement your data. The leveling stake is used to mount and level the LI-190R Quantum Sensor.
LI-200R Pyranometer and leveling stake
Part number 200R-8200
The LI-200R Pyranometer can be connected to a chamber to measure solar radiation at the chamber to supplement your data. The leveling stake is used to mount and level the LI-200R Pyranometer.
Stevens HydraProbe
Part number 900-19016
A Stevens HydraProbe soil moisture, temperature, and electrical conductivity probe with a cable connector is available to connect to the 8200-104/C Long-Term Chambers. Ancillary data from the probe is captured alongside soil gas flux data.
SDI-12 sensor cable with flying leads
Part number 392-18518
Optional 2-meter cable with connector and flying leads to connect an SDI-12 sensor to the 8200-104/C Long-Term Chambers.
Sensor cable T-split fitting
Part number 310-18516
The sensor T-split fitting connects directly to a chamber port and serves as a junction for the connectors from two sensors, such as a light sensor and chamber thermistor, or multiple SDI-12 sensors to the 8200-104/C Long-Term Chambers.
Flask sampling kit
Part number 8250-660
The flask sampling kit is an accessory that modifies the LI-8250 to measure fluxes from discrete samples, such as soil samples, fruits, or small animals in user-supplied chambers, referred to as flasks. When used for flask sampling, the LI-8250 is compatible with 8250-01 Extension Manifolds, for up to 36 flasks. For gas measurements it is compatible with LI-78xx Trace Gas Analyzers and the 8250-02 Calibration Manifold, supporting measurements of CO2, CH4, N2O, or Isotopes of CO2.
See Support: LI-8250 Multiplexer for theory and operating information.
| Description | Quantity | Part Number |
|---|---|---|
| Thermistor Input Module | 1 | 9982-079 |
| Purge pump | 1 | 9982-077 |
| Purge pump mount | 1 | 9982-042 |
| Bev-A-Line® tubing (15 m) | 2 | 8150-250 |
| M3-0.5 x 6 mm mounting screw | 10 | 150-14477 |
| M3-4.5 x 14 mm standoffs | 4 | 161-18848 |
| Washer | 20 | 167-07256 |
| Quick-connect bulkhead (sockets) | 8 | 300-07126 |
| Quick-connect bulkhead (pins) | 8 | 300-07127 |
| Barbed quick-connect plug (pins) | 16 | 300-07124 |
| Barbed quick-connect plug (sockets) | 16 | 300-07125 |
| Quick-connect plug | 1 | 300-08151 |
| Quick-connect fitting | 2 | 300-08251 |
| 3/32" hex wrench | 1 | 610-04290 |
Air Temperature Thermistor
Part number 9982-081
Preassembled, weatherized thermistors potted into a stainless-steel housing, fit to a sealed cable gland, and include a 15-m shielded cable. The cable gland mounts through a 12.5 mm diameter hole in materials up to 3 mm thick and is held in place by a nylon hex nut. The thermistor is compatible with the Thermistor Input Module and at least one should be used with flask configurations.
Custom chamber kit
Part number 8200-402
LI-COR offers a custom chamber kit that can be used to connect a custom chamber with the LI-8250 Multiplexer or 8250-01 Extension Manifold.
| Description | Quantity | Part Number |
|---|---|---|
| Cable with connector bare leads | 1 | 310-19090 |
| Cable assembly | 1 | 9982-056 |
| Quick-connect bulkhead (sockets) | 1 | 300-07126 |
| Quick-connect bulkhead (pins) | 1 | 300-07127 |
| Seal washer | 2 | 167-07256 |
Gas analyzer options
The LI-8250 Multiplexer is compatible with a variety of gas analyzers, including the LI-870 CO2/H2O Analyzer and LI-78xx Trace Gas Analyzers from LI-COR It is compatible with third-party gas analyzers as well. Below are brief descriptions of each of these analyzers and the components necessary to connect them to the multiplexer.
LI-870 CO2/H2O Analyzer
Part number LI-870
The LI-870 CO2/H2O Analyzer is a durable, weatherized gas analyzer that provides CO2 data to the LI-8250 Multiplexer for soil CO2 flux measurement.
The LI-870 measures CO2 in air at concentrations from 0 to 20,000 ppm using non-dispersive infrared gas analysis. Water vapor measurements are used in corrections to report CO2 dry mole fraction with high accuracy. Power to the LI-870 is supplied from the LI-8250 Multiplexer.
LI-870 and 8250-02 cable and tube assembly
Part number 9982-010
The cable assembly includes a 1.2-meter sealed USB-A-to-USB-B cable (part number 392-17654), a 1.2-meter IP68 sealed power cable (9982-008), and two pieces of Bev-a-line tubing with quick-connect fittings.
It provides power, communications, and gas flow between the LI-8250 Multiplexer and the LI-870 CO2/H2O Analyzer (see LI-870 cables and tubes). The same cable is used with the 8250-02 Calibration Manifold.
LI-78xx Trace Gas Analyzer
Part numbers
LI-7810,
LI-7815,
LI-7820, or LI-7825
LI-COR Trace Gas Analyzers are high-precision, high-stability, laser-based instruments that use Optical Feedback-Cavity Enhanced Absorption Spectroscopy (OF-CEAS) to measure gas concentrations in air.
LI-COR Trace Gas Analyzers provide data to the LI-8250 Multiplexer, which is used to compute flux results (see LI-78xx Trace Gas Analyzer cables and tubes). Up to three Trace Gas Analyzers can be attached to the LI-8250 at one time.
- CO2: LI-7810, LI-7815, LI-7825.
- CH4: LI-7810,
- N2O: LI-7820.
- CO2 Isotopologues: LI-7825.
LI-78xx cable and tube assembly
Part number 9982-011
To connect one LI-78xx Trace Gas Analyzer. The assembly includes a 2-meter Ethernet cable with two sealed RJ-45 connectors (part number 392-17793) and two pieces of Bev-a-line tubing with compression fittings. One is required for each trace gas analyzer.
T-split tube for multiple gas analyzers
Part number 9982-073
Pre-assembled T-split tubing for connecting multiple gas analyzers. The T-split tubing connects directly to the LI-8250 Multiplexer and serves as a junction for the air-in and air-out lines from the analyzers.
8250-01 Extension Manifold
Part number 8250-01
The multiplexer can support up to four extension manifolds, bringing the maximum number of chambers to 36. Each extension manifold includes a cable assembly (part number 9982-056; see Cable and tube assembly for chambers and extension manifolds), and a spares kit (part number 9982-072).
Because an extension manifold occupies a port, each 8250-01 Extension Manifold effectively adds another seven chambers to your system. The multiplexer can support four extension manifolds, which allows you use up to 36 chambers. Table 1‑3 shows how many extension manifolds are needed for each range of chambers.
What's what
Similar to the LI-8250, the 8250-01 Extension Manifold has a main connector panel and a ports connector panel.
Main connection panel
The 8250-01 Extension Manifold connects to the LI-8250 Multiplexer for power, communications, and the transfer of air.
- Power and Communication: Connects to the LI-8250 Multiplexer for power and communications.
- Air In: Connects to the air outlet of the multiplexer.
- Air Out: Connects to the air inlet of the multiplexer.
Port connection panel
An 8250-01 Extension Manifold connects to one port on the LI-8250 Multiplexer, and each extension manifold can connect up to eight chambers using the cable and tube assembly (part number 9982-056).
- Ports: The designated port number for the connected device. Each port has air in, air out, and a power and communication connector.
- Air In: Connects to the chamber air outlet.
- Air Out: Connects to the chamber air inlet.
- Power and Communication: Provides power to and communication to chambers.
Interior components
The internal components control the flow of air between the LI-8250 Multiplexer and chambers. The interior panel includes an input for powering the extension manifold and status indicators for the extension manifold.
- Pump: Provides a steady flow of air through the system.
- Port Valve Manifold: Controls the flow to and from each chamber.
- Ready LED: Steady when the 8250-01 Extension Manifold has booting up.
- Power LED: Illuminates when the 8250-01 Extension Manifold is receiving power.
- Alternate Power Input: For connecting to the indoor AC-to-DC power supply (part number 8250-772).
Note: The 8250-01 Extension Manifold only requires an alternate power source when using the purge pump for the Flask Sampling Kit (part number 8250-660).
Spares kit
Part number 9982-111
This kit includes additional components and spare parts for the LI-8250 Multiplexer and 8250-01 Extension Manifold.
8250-02 Calibration Manifold
Part number 8250-02
An accessory for the LI-8250, the 8250-02 Calibration Manifold provides automated baseline adjustments for connected gas analyzers. This is especially important with the LI-7825 CO2 Isotope/NH3 Trace Gas Analyzer, which requires daily baseline adjustments. It also features a patent-pending system to buffer transient changes to the water vapor concentration during a measurement. Physically reducing variation in water vapor reduces the need for mathematical corrections, thereby reducing uncertainty in measurements. This reduction in uncertainty is especially important when measuring low-abundance gas species such as CO2 isotopologues.
What's what
The 8250-02 Calibration Manifold features air inlets and outlets for gas analyzers and the multiplexer, and connector for gas standards.
Main panel
The main panel features a power connector and USB connector. Power, data, and control of the manifold is handled by the LI-8250 Multiplexer.
Gas standards connector panel
The gas standards panel has four compression fittings for 1/8" tubing. We recommend connecting copper or stainless steel tubing to these ports.
Spares and accessories kit
Part number 9982-106
The 8250-02 Calibration Manifold includes an accessories kit with the cable and tube assembly and 1/8" ferrules and nuts for the compression fittings.
| Description | Part Number |
|---|---|
| 1/8" Stainless Steel Ferrule and Nut Set | 300-21362 |
| Cable and Tube Assembly | 9982-010 |
LI-870 and 8250-02 cable and tube assembly
Part number 9982-010
The cable assembly includes a 1.2-meter sealed USB-A-to-USB-B cable (part number 392-17654), a 1.2-meter IP68 sealed power cable (9982-008), and two pieces of Bev-a-line tubing with quick-connect fittings.
SoilFluxPro™ Software
SoilFluxPro is a free Windows- and macOS-compatible application for advanced analysis of soil flux datasets. With SoilFluxPro, you can recompute datasets with altered parameters, view charts and summary statistics, and more. SoilFluxPro is available at the LI-COR technical support site (licor.com/support/LI-8250/software.html).
