Troubleshooting

Eventually, we all run into situations where things just do not go according to plan. This section will (hopefully) help you find a solution to some common issues. Should you run into an issue you cannot resolve, you can visit licor.com/env/support or you can reach LI-COR Technical Support at envsupport@licor.com.

System will not power on or powers on incompletely

  • Is the power supply strong enough? The system power supply should be capable of powering the entire system. Verify that your power supply provides enough current for all connected instruments. See Table F‑1 to see total system power requirements.
  • Are the cables properly tightened? Some of the threaded connections can appear to be connected but may require a little extra effort to fully seat them. Ensuring these cables are properly tightened is also important for achieving a weather-tight seal. Push in on the cable while giving it a little wiggle and turn the threaded nut and see if the system does not power on then.

Connection and communication problems

This section addresses problems communicating with the LI-8250 Multiplexer.

Cannot connect to the LI-8250 Multiplexer over Wi-Fi

If you are using Chrome, Firefox, or Safari, and are having trouble connecting, first try the steps outlined in Connecting over Wi-Fi. If you are still having trouble here are some potential solutions.

  • Has the system finished powering up? The system takes about a minute to fully power up. During this powering up period the Ready light on the interior panel will flash and the multiplexer will not broadcast a Wi-Fi signal. Once the Ready light is steady, you can try to connect.
  • Is Wi-Fi enabled? If Wi-Fi has been previously disabled for any reason you will need to re-enable it before connecting, or perhaps you did re-enable the Wi-Fi previously but did not press Apply. Instructions for re-enabling Wi-Fi can be found in Re-enabling Wi-Fi.
  • Have you entered the correct password? To connect to the LI-8250 Multiplexer Wi-Fi network enter the password licorenv to connect. Some devices may first prompt you to enter a PIN. In this case, select Use a security key instead.
  • You can try switching the Wi-Fi channel. This can be especially helpful where multiple wireless networks are available. To switch the Wi-Fi channel, you will first need to connect to the multiplexer user interface using an Ethernet connection (see Connecting over Ethernet). From the user interface, open the Wi-Fi settings and use the Channel drop-down menu to switch the Wi-Fi to another channel.
  • Using a Windows PC? We recommend using Windows 10, Version 1809 or higher. If you are using an older version, you may consider updating.

Cannot connect to the LI-8250 Multiplexer over Ethernet

If you are using Chrome, Firefox, or Safari, and are having trouble connecting, first try the steps outlined in Connecting over Ethernet. If you are still having trouble here are some potential solutions.

  • Has the system finished powering up? The system take about a minute to fully power up. During this powering up period the Ready light on the interior panel will flash and the multiplexer will not accept a connection. Once the Ready light is steady, you can try to connect.
  • Is your computer using Wi-Fi? You may need to disable Wi-Fi on your computer as the computer may default to this as the preferred connection.
  • Did you change networks? Network settings persist on the LI-8250 and require a restart before the settings will update. If you have changed networks, such as disconnecting the multiplexer from a LAN to connect via Ethernet cable, reboot the multiplexer and try connecting again.
  • You can also try to connect using the LI-8250 updater. Open the updater software, select your multiplexer from the list, and click Connect to LI-8250.

Cannot connect to the user interface

  • Ensure that you are connected to the LI-8250 Multiplexer Wi-Fi network. If you are somewhere where your device is set to automatically connect to a Wi-Fi network, the multiplexer connection may have been overridden and would require you to reconnect. You should also note that some devices will automatically disconnect themselves from a Wi-Fi network if they cannot access the Internet. This is an option that must be disabled on your device.
  • Make sure you have entered the correct serial number in the correct format in the browser address bar. See Connecting to the LI-8250 Multiplexer for details on where to find your serial number and how to enter that information into the address bar.
  • Are you using an Android device? The Android operating system does not support multicast Domain Name Service (mDNS), so you must use an alternative address to connect wirelessly with Android devices: http://82m-nnnn.li8250.licor.com.

A note on mDNS and IPV6

For speed and security, the LI-8250 Multiplexer was designed with the latest networking technology, which includes mDNS and IPV6. These technologies are not fully supported by all devices and operating systems. There are two main cases where mDNS and IPV6 limitations create LI-8250 Multiplexer connection issues:

  1. Android devices, which do not support mDNS.

Connecting to the multiplexer using the localhost requires mDNS, which is not currently supported by the Android OS. The fix for this is simple: use a different address: http://82m-nnnn.li-8250.licor.com, where 82m-nnnn is the serial number of your multiplexer.

  1. Windows operating systems.

Windows officially began supporting IPV6 with Windows 10 version 1809. You may encounter issues connecting via localhost to your LI-8250 Multiplexer if you are using an older version. In this case, we recommend updating your operating system, though this is not absolutely required for connection.

If you are not able, or choose not, to update your operating system, you may also connect using the multiplexer IP address. An easy way to identify your multiplexer IP address is through the software updater. Follow the instructions in Updating the firmware to download and install the updater. Connect your LI-8250 Multiplexer to your computer via Ethernet, and open the updater software. Click the serial number of your multiplexer, and the IP address will be displayed.

Write this address down for your records. It can be a useful backup if you run into issues connecting using the localhost. Enter the IP address directly into your browser to connect to the user interface.

If you cannot identify your multiplexer with the updater software, you should check to ensure the multiplexer is being recognized by your network. With your LI-8250 Multiplexer connected to your computer via Ethernet, open a command-line terminal and use the ping or nslookup commands, where 82m-nnnn is the serial number of your multiplexer.

  • ping 82m-nnnn.local (macOS)
  • nslookup 82m-nnnn.local (Windows)

If you receive a message like "Ping request could not find host 82m-nnnn.local. Please check the name and try again," your multiplexer is not being recognized by your network. In this case:

  • Is your Ethernet cable damaged or not connected properly?
  • Are you using the RJ-45 Ethernet cable supplied with your LI-8250 Multiplexer?
  • Is the Ethernet port and/or dongle of your computer working properly?
  • Are there any settings on your computer that would prevent you from connecting to devices via Ethernet?

LI-COR long-term chamber problems

See this section for issues related to the 8200-104 Opaque Long-Term Chamber and the 8200-104C Clear Long-Term Chamber.

Chamber will not open or close

  • Is there debris in the chamber? First check for and clear any debris that may be blocking the movement of the chamber.
  • Are the cable connections tight? Check that the OPEN/CLOSE button on the chamber is lit with a green light. If this light is not lit, check the cable connections to the PWR/COM port of the chamber and the port of the LI-8250 Multiplexer. Press in and tighten each connector or try to disconnect and reconnect the chamber, then try to open or close the chamber.
  • If the light around the OPEN/CLOSE button on the chamber is still not lit and you have checked the connections, try connecting the chamber to a different port on the multiplexer. You can also try a different cable assembly to see if the cable is damaged.
  • Try to open or close the chamber using the user interface. While connected to the LI-8250 Multiplexer user interface, go to Tools drop-down and select Manual System Control. Find the port number that the chamber is on and toggle between closed and open.

Chamber appears as orange in the Device Network

  • Has the chamber and all components connected to it been configured? If the chamber or its components have not been configured, it will appear in the device network as orange. See An overview of configurations for details.

Chamber appears as red in the Device Network

This indicates that the chamber and the LI-8250 Multiplexer are not communicating. First go to the Diagnostics page in the user interface. There may be a code or some indicator there as to why the instruments are not communicating.

  • Try disconnecting the power and communications cable from the chamber for 30 seconds and reconnecting.
  • Reboot the multiplexer by going to the Device Settings page and clicking Reboot at the bottom of the page.
  • Try connecting the chamber to a different port on the multiplexer.
  • If the problem persists, contact LI-COR Technical Support.

Chamber temperature readings are unreasonable

  • Is the thermistor connector fully seated in the sensor port? The thermistor connector may not be fully tightened. Try to fully tighten the thermistor connector and see if the problem persists.
  • Disconnect any other sensors from the chamber and see if the problem persists. An incorrect connection or short circuit in another sensor can cause a corrupted temperature measurement.
  • Have you tried connecting the thermistor to a different sensor port? If you are getting unreasonable readings from one port, try moving the thermistor to a different port and see if the problem persists. If the problem persists regardless of sensor port, the chamber thermistor may be damaged or malfunctioning. Contact LI-COR Technical Support.

Data and measurement problems

This section can help you address problems related to measurements and the quality of data.

Measurements stop immediately after starting

  • Is the power supply adequate? Be sure that your power supply is capable of running the entire system. If your power supply is not sufficient, the instruments will shut down when they attempt to draw a load. See Table F‑1 to see total system power requirements.

Data is noisy

  • Check the IN and OUT connections at the LI-8250 Multiplexer and the chamber to be sure the connections are complete. Also, inspect the tubing for any damage or leaks, especially near the connections.
  • Is the soil collar leaky? If the soil collar wobbles or if there are gaps between the soil and the bottom of the collar, there is likely a leak allowing ambient air into the chamber during a measurement. Reinstall the soil collar in another location. See Support: LI-8250 Multiplexer for how to install a soil collar.
  • Is the chamber sealing around the collar well? Look for any obstructions around the chamber and collar seals that may be interfering with a complete seal. If there are no obstructions, inspect the gaskets around the chamber and replace any damaged or hardened gaskets.

Initial gas concentrations are high

  • Is the prepurge long enough? One possible cause of a high initial gas concentration may be that the prepurge duration is not set long enough. Try to lengthen the prepurge and see if that corrects the problem. See Support: LI-8250 Multiplexer

Low/no flow during measurement

  • Is the filter on that port clogged? If the filter on that port has not been changed in a while, it may have become clogged and may need to be replaced. See Replacing the air filters to see how to change the filters. This may also be a good time to consider when you last changed the filter on the other ports or your car.
  • Is there a blockage or liquid water in the tube? Inspect the tubing for blockages, kinks, or liquid water in the tubing. If you can see a blockage, clean or replace the tubing using the spares kit as needed. If you see water, take off the tubing immediately and power off the LI-8250 Multiplexer. Try to get the water out as soon as possible, and consider contacting LI-COR Technical Support.
  • Has the pump gone bad? The pump on the LI-8250 Multiplexer and the 8250-01 Extension Manifold should run from the prepurge through the postpurge. Disconnect the tubing from the chamber and attempt an observation. Feel the ends of the tubing for air movement. If you do not feel air movement or if you cannot hear the pump, it may have failed. Contact LI-COR Technical Support.